Tek_isolved.pdf
Jak bezpiecznie zmierzyć napięcie sieciowe oscyloskopem SDS1052DL?
Warto zapoznać się jeszcze z tym: Link 1101446 1101453
Pobierz plik - link do postu
Fundamentals of Floating Measurements
and Isolated Input Oscilloscopes
Application Note
This guide will provide you with a fundamental glossary of
power measurement terms, explain the different options
available for making floating measurements, and highlight the
advantages and trade-offs of each option.
The most demanding floating measurement requirements are
found in power control circuits, such as motor controllers,
uninterruptible power supplies, and industrial equipment. In
such application areas, voltages and currents may be large
enough to present a hazard to users and/or test equipment.
When measuring floating high voltage signals there are many
options to consider. Each option has its advantages and
trade-offs.
Application Note
Differential vs. Floating Measurements
All voltage measurements are differential measurements. A
differential measurement is defined as the voltage difference
between two points. Voltage measurements fall into two
sub-categories:
1. round-referenced measurements
g
2. on-ground-referenced measurements (also known as
n
floating measurements)
Glossary
Common-Mode Signal
The component of an input signal which is common
(identical in amplitude and phase) on both inputs.
Common-Mode Range
The maximum voltage (from ground) of a common-mode
signal which a differential amplifier can reject.
Traditional Oscilloscopes
Common-Mode Rejection Ratio
Most traditional oscilloscopes have the “signal reference”
terminal connected to the protective grounding system,
commonly referred to as “earth” ground or just “ground”. This
is done so that all signals applied to, or supplied from, the
oscilloscope have a common connection point.
The performance measure of a differential amplifier’s ability
to reject common-mode signals. Because common-mode
rejection generally decreases with increasing frequency,
CMRR is usually specified at a particular frequency.
This common connection point is usually the oscilloscope
chassis and is held at (or near) zero volts by virtue of the thirdwire ground in the power cord for AC-powered equipment.
This means each input channel reference is tied to a single
ground reference.
The signal which is different between the two inputs of a
differential amplifier. The differential-mode signal (VDM) can
be expressed as:
A traditional passive probe should not be used to directly
make floating measurements on a ground referenced
oscilloscope. Depending on the amount of current flowing
through the reference lead, it can begin to get hot or, if the
current is high enough, it will melt open similar to the way a
fuse operates.
Differential Mode
VDM = (V+input)-(V-input)
Differential-Mode Signal
The signal which is different between two inputs.
Differential Measurement
The voltage difference between two points.
Floating Measurement Techniques
Differential Probe
The different options available for making high voltage floating
measurements are:
Voltage Isolators
A probe designed specifically for differential applications.
Active differential probes contain a differential amplifier at
the probe tip. Passive differential probes are used with
differential amplifiers and can be calibrated for precisely
matching the DC and AC attenuation in both signal paths
(including the reference lead).
“A minus B” Measurement Techniques
“Floating” the Oscilloscope Techniques
A differential measurement where neither point is referenced
to ground (earth potential).
Isolated-input Oscilloscopes
Differential Probes
Ground Loops
A ground loop results when two or more separate ground
paths are tied together at two or more points. The result is
a loop of conductor. In the presence of a varying magnetic
field, this loop becomes the secondary of a transformer
which acts as a shorted turn. The magnetic field which
excites the transformer can be created by any conductor
2
www.tektronix.com/tps2000b
Fundamentals of Floating Measurements and Isolated Input Oscilloscopes
Glossary (continued)
in the vicinity which is carrying a non-DC current. AC line
voltage in main wiring or even the output lead of a digital IC
can produce this excitation. The current circulating in the loop
develops a voltage across any impedance within the loop.
Thus, at any given instant in time, various points within a
ground loop will not be at the same AC potential.
Most bench-top oscilloscopes share the architecture shown
below. With this architecture, all input signals must have
the same voltage reference when taking multi-channel
measurements and the shared default reference is “earth”
ground. Without differential preamplifiers or external signal
isolators, these bench-top oscilloscopes are not suitable for
taking floating measurements.
h
1
Oscilloscope channels
must float together.
C
Connecting the ground lead of an oscilloscope probe to the
ground in the circuit-under-test results in a ground loop if
the circuit is “grounded” to earth ground. A voltage potential
is developed in the probe ground path resulting from the
circulating current acting on the impedance within the path.
Shared Reference vs. Isolated Channel
Architecture
C
Thus, the “ground” potential at the oscilloscope’s input
BNC connector is not the same as the ground in the circuit
being measured (i.e., “ground is not ground”). This potential
difference can range from microvolts to as high as hundreds
of millivolts. Because the oscilloscope references the
measurement from the shell of the input BNC connector, the
displayed waveform may not represent the real signal at the
probe input. The error becomes more pronounced as the
amplitude of the signal being measured decreases.
h
2
C
h
3
n
ai is
M ass
h
C
C
h
4
Case
device under test
power
supply
Connecting the ground lead of an oscilloscope probe to the ground in the circuitunder-test results in a ground loop if the circuit is “grounded” to earth ground.
“A Single”
0607-014
Compared to the conventional bench oscilloscope
architecture, voltage references in the isolated channel
architecture are not connected together inside the
instrument. Therefore, each reference of the used inputs
must be connected to a reference voltage. Independently
floating isolated inputs are still coupled by parasitic
capacitance. This can occur between the input references
and the environment, and between the input references
mutually. For this reason, it is advisable to connect references
to a system ground or another stable voltage. If the reference
of an input is connected to a high speed and/or high voltage
signal, you should be aware of parasitic capacitance.
C
h
h
C
4
1
Oscilloscope channels and
Comm Interface
float independently.
DC and / or
AC voltage
h
C
2
h
C
n
ai is
M ass
h
C
3
m e
om c
C erfa
t
In
Battery-operated oscilloscopes with grounded input channels
when operated from AC line power and using a standard 3
wire power cord, exhibit the same limitations as traditional
oscilloscopes. However, when operating on battery power,
these oscilloscopes allow you to make a single, safe floating
measurement up to 30VRMS at a time. Remember that all
input commons are tied together.
DC and / or
AC voltage
Electrically insulated case
3
Application Note
Description
Isolated Input
Oscilloscopes
Measurements
Oscilloscopes with IsolatedChannel™
input architecture, such as the
TPS2000B or THS3000 Series, provide
true and complete channel-to-channel
and channel-to-power line isolation.
Each channel is individually isolated
from one another and other
non-isolated components. When
making floating measurements with an
IsolatedChannel™ oscilloscope, one
must use specifically designed passive
probes, such as the TPP0201 to float
up to 30 VRMS, the THP0301 to float up
to 300 VRMS or the P5122/P5150
probes to float up to 600 VRMS. Unlike
the passive probes used with most
Differential
Probe
Measurements
Differential probe systems enable
floating measurements to be made
with the Tektronix TDS/DPO/MSO and
most other grounded oscilloscopes.
Some differential probes such as the
P6246, P6247, P6248, and P6330 are
optimized for fast, lower amplitude
signals. Others such as the P5200A,
P5205A, and P5210A handle slower
signals with higher voltage amplitudes.
The ADA400A differential preamplifier
provides the capability to display
low frequency, very low amplitude
differential signals even in high noise
environments.
Voltage Isolator
Measurements
As the name implies, isolators do not
have direct electrical connection
between the floating inputs and their
ground-referenced outputs. The
signal is coupled via optical or
split-path optical/transformer means.
“A minus B”
Measurements
The “A minus B” measurement
technique allows the use of a
conventional oscilloscope and its
passive voltage probes to indirectly
make floating measurements. One
channel measures the “positive”
test point and another channel
measures the “negative” test point.
Subtracting the second from first
removes the voltage common to
both test points in order to view the
floating voltage that could not be
measured directly. Oscilloscope
channels must be set to the same
volts/division; the probes should
be matched to maximize common
mode rejection ratio.
(Also known as the
Pseudo-Differential
Measurements)
“Floating” a
Conventional
Grounded
Oscilloscope
Traditional Ground Reference Oscilloscope
Channel 2
Probe
VCG
CH1
CH2
CH3
CH4
CH5
Channel 1
Probe
VCG = Channel 2 - Channel 1
Example of two probes measuring ground reference voltages.
A common but risky practice is to float
the oscilloscope through the use of
an isolation transformer that does
not carry the ground through to the
secondary or by disconnecting the
oscilloscope’s AC mains power cord
grounding connector.
WARN I N G
This is an unsafe and dangerous
practice and should never be done!
Failure to follow safety warnings can
result in serious injury or loss of life.
Curre
n
t
“Floating” a ground referenced
oscilloscope puts all accessible
metal including the chassis, casing,
and connectors at the same
voltage as the test point that the
probe reference lead is connected.
conventional oscilloscopes, these
types of probes are insulated at the
BNC connection for shock protection
and the reference lead is designed to
withstand the rated float voltage. (For
more information please refer to the
discussion entitled “Mind your CATs
and Volts” later in this application note).
UUT
VMeas
V1
Scope Chassis =
V1
fe!
Unsa
Earth Ground
A floating measurement in which dangerous voltages occur on the oscillioscope chassis.
V1 may be hunderouds of volts!
4
Advantages & Trade-Offs
Advantages
Isolated input channel oscilloscopes
offer a safe and reliable way to make
floating measurements. An obvious
benefit of channel-to-channel and
channel-to-ground isolation is the
ability to simultaneously view multiple
signals referenced to different voltages.
Another benefit is the ability to do this
without the added cost of specialized
probes or expensive and bulky voltage
isolators. Channel-to-power line
isolation eliminates the path between
ground of the signal source and
the oscilloscope.
Trade-Offs
Unlike a differential probe, an isolated
input channel does not provide a
balanced floating measurement. The
impedance to earth ground is different
in the tip (+) input and reference (-)
input. Since the reference (-) input of
the isolated channel does not have a
default reference level like a grounded
oscilloscope, you must connect the
Advantages
Differential probes provide a safe
method to adapt a grounded
oscilloscope to make floating
measurements. In addition to the safety
benefits, the use of these probes can
improve measurement quality.
Differential probes provide balanced
measurement input capacitance so any
point in the circuit can be safely probed
with either lead. Differential probes
also typically have better CMRR
performance at higher frequencies
than voltage isolators.
Trade-Offs
The probes still have a resistive path
to ground - so if a circuit is sensitive
to leakage currents then differential
probes may not be the best solution.
Advantages
Voltage isolators provide a means of
safely measuring floating voltages.
Because isolators have no resistive
path to ground, they are a good choice
for applications which are extremely
sensitive to leakage currents.
Trade-Offs
Voltage isolators add a layer of cost.
An independent power supply and
isolation amplifier box must
Advantages
The advantage of of using the " A minus
B " measurement technique is that it
can be easily done with almost any
oscilloscope and its standard probes.
Keep in mind that both test points
must be referenced to ground. Thus
this method will not work if either test
point is floating or if the entire system is
floating.
Trade-Offs
Two oscilloscope channels are used to
make one “A minus B” measurement.
The primary limitation of this technique
is the rather small common-mode
range, which results from the oscilloscope’s vertical channel dynamic
range. Generally this is less than ten
times the volts/division setting from
ground. Whenever common mode
voltage is greater than differential
mode voltage, the “A minus B”
technique can be thought of as
extracting the small differences from
two large voltages. This technique is
suitable for applications where the
common-mode signal is the same or
lower amplitude than the differentialmode signal, and the common-mode
component is DC or low frequency,
such as 50 or 60 Hz power line. It
effectively eliminates ground loop
voltages from the measurement
when measuring signals of moderate
amplitude.
Advantages
Although floating equipment is a
method that uses existing equipment
to make floating measurements and
remove ground loops on lower
frequency signals, it is an unsafe and
dangerous practice and should never
be done.
immediate failure but can lead to future
dangerous failures (a shock and fire
hazard), even after returning the
oscilloscope to properly grounded
operation.
Any source impedance the reference
is connected to will be loaded during
fast common-mode transi- tions,
attenuating the signal. Worse yet, the
high capacitance can damage some
circuits. Connecting the oscilloscope
common to the upper gate in an
inverter may slow the gate-drive signal,
preventing the device from turning
off and destroying the input bridge.
This failure is usually accompanied
by a miniature fireworks display right
on your bench.
Trade-Offs
This technique is dangerous, not
only from the standpoint of elevated
voltage present on the oscilloscope
(a shock hazard to the operator), but
also due to cumulative stresses on
the oscilloscope’s power transformer
insulation. This stress may not cause
Another benefit is the full use of the
oscilloscope’s multiple channels
with the simultaneous viewing of
multiple signals, referenced to
different voltages.
At higher frequencies, severing the
ground may not break the ground
loop as the line-powered instrument
exhibits a large parasitic capacitance
when floated above earth ground.
The floating measurement can be
corrupted by ringing. Floating
oscilloscopes do not have balanced
inputs. The reference side (the
“ground” clip on the probe) has a
significant capacitance to ground.
reference lead of the probe to the
reference on the DUT.
There is no shunting to ground,
therefore line frequency fields radiated
from fluorescent lighting and building
wiring can result in more baseline
noise on the oscilloscope reading.
Using averaging acquisition mode can
mitigate this increase in baseline noise.
cost and bulk. The gain and offset
characteristics must be factored
manually into every measurement.
Other trade-offs include an added
layer of cost - depending on the
oscilloscope capability an independent
power supply may be required, adding
Yet another trade-off is that only one
measurement may be made at a time remember all the input references are
be used. The gain and offset
Actual
Signal
“Ringing”
Ringing caused by parasitic inductance
and capacitance distorts the signal and
invalidates measurements.
tied to each other. Once you have
floated one input references, all
input references are now floating at
the same level.
5
Application Note
Tektronix TPS2000B and THS3000
Series IsolatedChannel™ Oscilloscopes
The TPS2000B Series combines the proven Tektronix bench
oscilloscope performance with a four isolated channel
architecture designed for making measurements on products
operating on industrial power. Paired with the optional power
bundle (TPS2PBND2), which includes four passive, high
voltage probes (P5122) and the power measurement and
analysis software package (TPS2PWR1), this instruments sets
the standard for advanced floating measurements in its class.
The power measurement and analysis software package
offers power analysis measurements (true power, reactive
power, true power factor, phase angle), waveform analysis
measurements (RMS, crest factor, frequency), harmonic
measurements, and switching loss measurements.
The THS3000 Series is designed for engineers requiring a
higher degree of mobility in their oscilloscope usage without
sacrificing bench-level performance or the ability to perform
floating measurements. The instrument’s ruggedized,
light-weight and handheld form factor, 7hr single-battery
performance and automated measurement capability offers a
large degree of versatility in both lab and field operations. The
instrument’s isolated channels, high voltage acquisition (up to
1000 Vrms CAT II) and advanced waveform recording capability
allows for safe floating measurements and is therefore
highly suitable for users who frequently need to correlate
measurement taken in the lab to those taken in the field.
Mind your CATs and Volts
Summary Description of Categories
CAT IV
For measurements performed at the source of a low voltage installation
( & lt; 1,000 V).
CAT III
For measurements performed in the building installation.
How to select probe and oscilloscope combination:
CAT II
For measurements performed on circuits directly connected to the low
voltage installation.
1. Determine measurement (or overvoltage) category
CAT I
For measurements performed on circuits not directly connected to MAINS.
Selecting the proper voltage probe
The IEC 61010-1 international standard defines four
overvoltage categories for voltage-measuring instruments.
The overvoltage categories I through IV are defined by how
much electrical energy could be present during a transient.
Based on IEC 61010-1, voltage-measuring instruments
are rated on their ability to withstand a voltage transient.
3. Determine maximum tip to ground voltage.
4. Determine the maximum voltage from probe tip to
the reference lead.
5. Determine maximum peak to peak reading desired
on screen.
2. Determine maximum floating voltage.
transmission lines
transformer
circuit breaker
Uncontrolled
Category IV
Primary Supply Level
IEC Installation Categories.
6
isolating
device
outlet
Category III
Category II
Category I
Fixed Installation
Appliances and
Portable Equipment
Telecommunication and
Electronic Equipment
Mind your CATs and Volts (continued):
Selecting the proper voltage probe for your
TPS2000B and THS3000 Series Oscilloscopes
Probe Name
TPP0201
THP0301
P5150
P5122*1
Maximum probe tip to
ground voltage
300 VRMS
CAT II
300 VRMS
CAT III
1000 VRMS
CAT II
1000 VRMS
CAT II when
DC coupled
Maximum reference to
ground (floating) voltage
30 VRMS
300 VRMS
CAT III
600 VRMS
CAT II
600 VRMS
CAT II
Attenuation settings
10x
10x
50x
100x
Bandwidth
200 Mhz
300 Mhz
500 Mhz
200 Mhz
Probe type
passive
passive
passive
passive
On-screen Peakto-Peak Voltage2
TPS
400 VP-P
400 VP-P
2000 VP-P
2828 VP-P
THS
849 VP-P
849 VP-P
2828 VP-P
2828 VP-P
On-screen RMS
Voltage2
TPS
141 VRMS
141 VRMS
707 VRMS
1000 VRMS
THS
300 VRMS
300 VRMS
1000 VRMS
1000 VRMS
P5122 probe should not be used for AC-coupled measurements on signals with DC & gt; 300 V
on the TPS2000
2
Limited by instrument's vertical resolution (TPS = 5V/div, THS = 100V/div)
1
High-voltage
Oscilloscope Probe
Oscilloscope Input BNC
CH 1
Input Signal Rating
Reference Float Rating
2. Determine maximum floating voltage required
(ground to reference voltage)
This example has 240 VRMS measurement requirements.
Therefore, only the THP0301, P5150 and P5122
probes would be suitable.
3. Determine maximum probe tip to ground voltage
requirements
In this example, the engineer determined this to be
approximately 140 VRMS.
All four probes fulfill this requirement.
4. Determine maximum voltage from probe tip to the
reference lead
The maximum expected voltage is the 240 VRMS of the
signal to be measured.
This limits the probe selection to the THP0301, P5150
and P5122 probes.
5. Determine attenuation needed
Calculate maximum peak-to-peak voltage
corresponding to 240 VRMS.
Desired viewable voltage range is therefore:
240 VRMS x √2 x 2 = 679 VP-P.
TPS2000B Series has maximum vertical setting of
5V/div and 8 divisions on the screen.
THS3000 Series has maximum vertical setting of
100V/div and 8 divisions on the screen.
Probe attenuation needed = viewing voltage / max
vertical / # of divisions.
Example:
Need to measure peak-to-peak voltages on Line-to-Line 240
VRMS 3-phase wye Active Harmonic Filter.
1. Determine maximum rated input voltage for desired
installation category: CAT III
The maximum allowed input voltage for a certain CAT
rating generally applies to lower CAT environments as
well. E.g., if the instrument is rated 300 VRMS CAT III, it
would be safe to work with the instrument at 300 VRMS
in a CAT II environment.
TPS2000B: 679 VP-P / 5 / 8 = 17x, only the P5150 and
P5122 offer enough attenuation.
THS3000: 679 VP-P / 100 / 8 = 1x, all available probes
6. Select probe fulfilling minimal requirements
For the TPS2000B,the P5150 and P5122 would be
suitable probes for this test scenario
For the THS3000, the THP0301, P5150 and P5122
would be suitable probes.
Instruments are often additionally rated at a lower CAT
category for higher voltage levels, e.g. 300 VRMS CAT III
and 600 VRMS CAT II simultaneously. This is not always
the case and therefore such ratings across multiple CAT
categories must be explicitly verified by the user in the
instrument’s documentation.
7
Contact Tektronix:
ASEAN / Australasia (65) 6356 3900
Austria* 00800 2255 4835
Balkans, Israel, South Africa and other ISE Countries +41 52 675 3777
Belgium* 00800 2255 4835
Brazil +55 (11) 3759 7627
Canada 1 (800) 833-9200
Central East Europe and the Baltics +41 52 675 3777
Central Europe & Greece +41 52 675 3777
Denmark +45 80 88 1401
Finland +41 52 675 3777
France* 00800 2255 4835
Germany* 00800 2255 4835
Hong Kong 400-820-5835
India 000-800-650-1835
Italy* 00800 2255 4835
Japan 81 (3) 6714-3010
Luxembourg +41 52 675 3777
Mexico, Central/South America & Caribbean 52 (55) 56 04 50 90
Middle East, Asia and North Africa +41 52 675 3777
The Netherlands* 00800 2255 4835
Norway 800 16098
People’s Republic of China 400-820-5835
Poland +41 52 675 3777
Portugal 80 08 12370
Republic of Korea 001-800-8255-2835
Russia & CIS +7 (495) 7484900
South Africa +27 11 206 8360
Spain* 00800 2255 4835
Sweden* 00800 2255 4835
Switzerland* 00800 2255 4835
Taiwan 886 (2) 2722-9622
United Kingdom & Ireland* 00800 2255 4835
USA 1 (800) 833-9200
* If the European phone number above is not accessible,
please call +41 52 675 3777
Contact List Updated 10 February 2011
For Further Information
Tektronix maintains a comprehensive, constantly expanding collection of
application notes, technical briefs and other resources to help engineers
working on the cutting edge of technology. Please visit www.tektronix.com
Copyright © 2011, Tektronix. All rights reserved. Tektronix products are
covered by U.S. and foreign patents, issued and pending. Information in this
publication supersedes that in all previously published material. Specification
and price change privileges reserved. TEKTRONIX and TEK are registered
trademarks of Tektronix, Inc. All other trade names referenced are the service
marks, trademarks or registered trademarks of their respective companies.
11/11
EA/FCA-POD