UT60AEngManualLVD.zip
UNI-T UT60A instrukcja obsługi ENGLISH
UNI-T UT60A instrukcja obsługi ENGLISH Producent: UNI-T Model: UT60A Jezyk: ENGLISH
Pobierz plik - link do postu
  • UT60AEngManualLVD.zip
    • UT60AEngManualLVD.pdf
UT60AEngManualLVD.zip > UT60AEngManualLVD.pdf
Model UT60A: OPERATING MANUAL
Table of Contents
Title
Page
Overview
3
Unpacking Inspection
4
Safety Information
5
Rules For Safe Operation
6
International Electrical Symbols
8
The Meter Structure
9
Rotary Switch
10
Functional Buttons
11
Display Symbols
14
Measurement Ranges
17
A.
Selecting a Measurement Range
17
B.
Manual Ranging and Autoranging
17
Measurement Operation
19
A.
AC & DC Voltage Measurement
19
B.
Measuring Continuity, Diodes &
Resistance
21
C.
Capacitance Measurement
24
D.
Frequency & Duty Cycle Measurement
26
E.
DC or AC Current Measurement
28
The POWER Button
30
The BLUE Button
30
The Use of Relative Value Mode
30
Operation of Hold Mode
31
General Specifications
32
Accuracy Specifications
33
A.
AC Voltage
33
B.
DC Voltage
33
C.
Continuity, Diodes & Resistance Test
34
D.
Capacitance
35
E.
35
F.
DC Current
36
G.
AC Current
36
1
Title
Maintenance
A.
General Service
B.
Testing the Fuses
C.
Replacing the Battery
D.
Replacing the Fuses
RS232C Serial Port
A.
RS232C Port Cable
B.
Setting of RS232C Serial Ports
C.
System Requirements for Installing
the UT60A Interface Program
2
Page
37
37
37
38
39
41
41
41
42
Overview
This Operating Manual covers information on safety and
cautions. Please read the relevant information carefully
and observe all the Warnings and Notes strictly.
Warning
To avoid electric shock or personal injury, read the
" Safety Information " and " Rules for Safe Operation "
carefully before using the Meter.
Digital Multimeter Model UT60A (hereafter referred to
as " the Meter " ) has autorange and manual range options
with maximum reading 3999. The enclosure structure
design adopted advanced " co-injection " technique in
order to provide sufficient insulation.
In addition to the conventional measuring functions, it
is equipped with a RS232C standard serial port for easy
connection with computer to realize macro recording and
monitoring and capture of transient dynamic data,
displaying change of waveform during the measurement,
providing data and evidence to engineering technicians
for scientific research. This is also a highly applied digital
multimeter of good performance with full overload
protection.
3
Open the package case and take out the Meter. Check
the following items carefully to see any missing or
damaged part:
Description
Qty
Operating Manual
1 piece
Test Lead
1 pair
9V Battery (NEDA1604, 6F22 or 006P) 1 piece
(installed)
4 RS232C Interface Cable
1 piece
5 CD-ROM (Installation Guide & Computer 1 piece
Interface Software)
In the event you find any missing or damage, please
contact your dealer immediately.
Item
1
2
3
4
Safety Information
This Meter complies with the standards IEC61010: in
pollution degree 2, overvoltage category (CAT. III 1000V,
CAT. IV 600V) and double insulation.
CAT. III: Distribution level, fixed installation, with smaller
transient overvoltages than CAT. IV
CAT IV:Primary supply level, overhead lines, cable
systems etc.
Use the Meter only as specified in this operating manual,
otherwise the protection provided by the Meter may be
impaired.
In this manual, a Warning identifies conditions and
actions that pose hazards to the user, or may damage
the Meter or the equipment under test.
A Note identifies the information that user should pay
attention on.
International electrical symbols used on the Meter and
in this Operating Manual are explained on page 8.
5
Rules For Safe Operation (1)
Warning
To avoid possible electric shock or personal injury,
and to avoid possible damage to the Meter or to the
equipment under test, adhere to the following rules:
l
l
l
l
l
l
l
l
l
Before using the Meter inspect the case. Do not
use the Meter if it is damaged or the case (or
part of the case) is removed. Look for cracks
or missing plastic. Pay attention to the insulation
around the connectors.
Inspect the test leads for damaged insulation or
exposed metal. Check the test leads for
continuity. Replace damaged test leads with
identical model number or electrical
specifications before using the Meter.
Do not apply more than the rated voltage, as
marked on the Meter, between the terminals or
between any terminal and grounding.
The rotary switch should be placed in the right
position and no any changeover of range shall
be made during measurement is conducted to
prevent damage of the Meter.
Never working at an effective voltage over 60V
in DC or 30V rms in AC for there is danger of
electric shock.
Use the proper terminals, function, and range
for your measurements.
Do not use or store the Meter in an environment
of high temperature, humidity, explosive,
inflammable and strong magnetic field. The
performance of the Meter may deteriorate after
dampened.
When using the test leads, keep your fingers
behind the finger guards.
Disconnect circuit power and discharge all highvoltage capacitors before testing continuity,
diodes, resistance, capacitance or current.
6
Rules For Safe Operation (2)
l
l
l
l
l
l
l
l
l
l
Before measuring current, check the Meter's
fuses and turn off power to the circuit before
connecting the Meter to the circuit.
Replace the battery as soon as the battery
indicator appears. With a low battery, the Meter
might produce false readings that can lead to
electric shock and personal injury.
Remove test leads and RS232C interface cable
from the Meter and turn the Meter power off
before opening the Meter case.
When servicing the Meter, use only the same
model number or identical electrical
specifications replacement parts.
The internal circuit of the Meter shall not be
altered at will to avoid damage of the Meter and
any accident.
Soft cloth and mild detergent should be used to
clean the surface of the Meter when servicing.
No abrasive and solvent should be used to
prevent the surface of the Meter from corrosion,
damage and accident.
The Meter is suitable for indoor use.
Under the environment with high (+/-4kV)
electrostatic discharge, the Meter may not be
operated as normal condition. The user may
require resetting the Meter.
Turn the Meter off when it is not in use and take
out the battery when not using for a long time.
Constantly check the battery as it may leak when
it has been using for some time, replace the
battery as soon as leaking appears. A leaking
battery will damage the Meter.
7
AC (Alternating Current).
DC (Direct Current).
AC or DC.
Grounding.
Double Insulated.
Deficiency of Built-In Battery.
Continuity Test.
Diode.
Capacitance Test.
Fuse.
Warning. Refer to the Operating Manual.
Conforms to Standards of European Union.
8
The Meter Structure
9
10
11
12
13
Display Symbols (1)
2
3
4
5
6
7
8
15
9
1
kM
nFHz
VA
AC
16
14
10 ~ 14
Display Symbols (2)
15
16
Measurement Ranges (1)
A measurement range determines the highest value the
Meter can measure. Most Meter functions have more
than one range. See " Accuracy Specifications. "
A. Selecting a Measurement Range
Being in the right measurement range is important:
l
l
If the range is too low for the input, the Meter displays
to indicate an overload.
If the range is too high, the Meter will not display
the most accurate measurement.
B. Manual Ranging and Autoranging
The Meter has both manual range and autorange options:
l
l
In the autorange mode, the Meter selects the best
range for the input detected.
This allows you to switch test points without having
to reset the range.
In the manual range mode, you select the range.
This allows you to override autorange and lock the
Meter in a specific range.
The Meter defaults to the autorange mode in
measurement functions that have more than one range.
When the Meter is in the autorange mode,
is
displayed.
17
Measurement Ranges (2)
To enter and exit the manual range mode:
1.
2.
Press RANGE.
The Meter enters the manual range mode and
turns off.
Each presses of RANGE increments the range.
When the highest range is reached, the Meter wraps
to the lowest range.
To exit the manual range mode, press and hold
RANGE for over 2 seconds.
The Meter returns to the autorange mode and
is displayed.
Note
l If you manually change the measurement range
after entering the REL or Hold modes, the Meter
exits these modes.
l Under frequency/duty cycle and capacitance
measurement, only autorange mode is available.
18
Measurement Operation (1)
A. AC & DC Voltage Measurement
Warning
To avoid harms to you or damages to the Meter from
electric shock, please do not attempt to measure
voltages higher than 1000V / 750V rms although
readings may be obtained.
AC Voltage Measurement (see figure 3)
Red
Black
(figure 3)
The AC voltage ranges are: 4.000V, 40.00V, 400.0V and
750.0V. To measure AC Voltage, connect the Meter as
follows:
1.
2.
3.
Insert the red test lead into the HzV terminal and
the black test lead into the COM terminal.
Set the rotary switch to Hz V and press BLUE
button to select AC measurement mode.
Connect the test leads across with the object being
measured.
The measured value shows on the display, which
is effective value of sine wave (mean value
response).
19
Measurement Operation (2)
Note
l In each range, the Meter has an input impedance
of 10M . This loading effect can cause measurement
errors in high impedance circuits. If the circuit
impedance is less than or equal to 10k , the error
is negligible (0.1% or less).
l When AC voltage measurement has been
completed, disconnect the connection between the
testing leads and the circuit under test.
DC Voltage Measurement (see figure 4)
Select
Hz
V
red
black
(figure 4)
The DC Voltage ranges are: 400.0mV, 4.000V, 40.00V,
400.0V and 1000V. To measure DC voltage, connect
the Meter as follows:
1.
2.
3.
Set the rotary switch to Hz V; DC measurement
is default or press BLUE button to select DC
measurement mode.
measured.
The measured value shows on the display.
20
Measurement Operation (3)
Note
l When DC voltage measurement has been
B. Measuring Continuity, Diodes & Resistance
Warning
To avoid damages to the Meter or to the devices
under test, disconnect circuit power and discharge
all the high-voltage capacitors before measuring
continuity, diodes & resistance.
Testing for Continuity (see figure 5)
Select
red
black
(figure 5)
To test for continuity, connect the Meter as below:
1. Insert the red test lead into the HzV terminal and
2. Set the rotary switch to
and press BLUE
button to select measurement mode.
3. The buzzer sounds if the resistance of a circuit under
test is less than 100 .
21
Measurement Operation (4)
Note
l The LCD displays
indicating the circuit being
tested is open.
l When continuity testing has been completed,
disconnect the connection between the testing leads
Testing Diodes (see figure 6)
Select
red
black
(figure 6)
Use the diode test to check diodes, transistors, and other
semiconductor devices. The diode test sends a current
through the semiconductor junction, and then measures
the voltage drop across the junction. A good silicon
junction drops between 0.5V and 0.8V.
To test a diode out of a circuit, connect the Meter as
follows:
1.
2.
3.
and press BLUE
button to select
measurement mode.
For forward voltage drop readings on any
semiconductor component, place the red test lead
on the component's anode and place the black test
lead on the component's cathode.
22
Measurement Operation (5)
Note
l In a circuit, a good diode should still produce a
forward voltage drop reading of 0.5V to 0.8V;
however, the reverse voltage drop reading can vary
depending on the resistance of other pathways
between the probe tips.
l Connect the test leads to the proper terminals as
said above to avoid error display.
l The LCD will display
indicating open-circuit for
wrong connection.
l The unit of diode is Volt (V), displaying the positiveconnection voltage-drop value.
l When diode testing has been completed, disconnect
the connection between the testing leads and the
circuit under test.
Resistance Measurement (see figure 7)
Select
red
black
(figure 7)
The resistance ranges are: 400.0 , 4.000k , 40.00k ,
400.0k , 4.000M
and 40.00M
. To measure
resistance, connect the Meter as follows:
1.
2.
, resistance
measurement ( ) is defaults or press BLUE button
23
Measurement Operation (6)
3.
measured.
Note
l The test leads can add 0.1 to 0.2 of error to
resistance measurement. To obtain precision
readings in low-resistance measurement, that is the
range of 400.0 , short-circuit the input terminals
beforehand, using the relative value function button
REL to automatically subtract the value measured
when the testing leads are short-circuited from the
reading.
l For high-resistance measurement ( & gt; 1M ), it is
normal taking several seconds to obtain a stable
reading.
l If reading with shorted test leads is not 0.5 ,
check for loose test leads, incorrect function selection,
or enabled Data Hold function.
l The LCD displays
indicating open-circuit for the
tested resistor or the resistor value is higher than
the maximum range of the Meter.
l When resistance measurement has been completed,
C. Capacitance Measurement (see figure 8)
red
black
(figure 8)
24
Measurement Operation (7)
Warning
To avoid damage to the Meter or to the equipment
all high-voltage capacitors before measuring
capacitance.
Use the DC Voltage function to confirm that the
capacitor is discharged. Never attempt to input over
60V in DC or 30V rms in AC to avoid personal
dangerous.
The Meter's capacitance ranges are: 40.00nF, 400.0nF,
4.000µF, 40.00µF, and 100.0µF. To measure capacitance,
1.
2.
3.
Insert the red test lead into the HzVΩ terminal and
Set the rotary switch to .
measured.
Note
l For testing the capacitor with polarity, connect the
red test lead to anode & black test lead to cathode
instead.
l To increase the accuracy of capacitance
measurement especially when measuring under
400nF, use REL mode to automatically subtract the
Meter built-in equalized capacitance and residual
capacitance of test leads from the result.
l It takes a longer time when testing a high capacitor
value, the testing time is around 15 seconds in
100µF range.
l When capacitance measurement has been
25
Measurement Operation (8)
D. Frequency & Duty Cycle Measurement (see figure 9)
Select
Hz Test
Red
Black
(figure 9)
Frequency Measurement
The measurement ranges are from 10Hz to 40MHz. To
measure frequency, connect the Meter as follows:
1.
2.
3.
Set the rotary switch to Hz%; frequency
measurement (Hz) is default or press Hz% button
to select Hz measurement mode.
measured.
Note
l To obtain a stable reading when measuring:
Input scope & gt; 30V rms & 1 kHz frequency signal:
Set the rotary switch to Hz V, A Hz,
mA Hz or Hz A.
Then press Hz% to select Hz measurement
mode to obtain frequency value.
Input scope 30V rms frequency signal:
Follow the above step 2 carrying out the
measurement.
l When making frequency measurement at voltage
or current range, please mind the following signal
requirement table:
26
Measurement Operation (9)
Range
l
Signal
Frequency
Requirement
Range
200mV
µA
200µA
10Hz~1kHz
mA
20mA
A
2A
When Hz measurement has been completed,
The duty cycle measurement range is 0.1% ~ 99.9%.
To measure duty cycle:
1. Set up the Meter to measure frequency.
2. To select duty cycle, press Hz again (or until the %
symbol is shown on the display).
3. Connect the test leads across with the object being
measured.
Note
l The LCD displays 000.0% indicating the input signal
is high or low level.
mA Hz or Hz A.
measurement.
l When duty cycle measurement has been completed,
27
Measurement Operation (10)
E. DC or AC Current Measurement (see figure 10)
Select DC
or AC
µA Hz,
mA Hz or
Hz A.
Black
Red
(figure 10)
Warning
Never attempt an in-circuit current measurement
where the open-circuit voltage between the circuit
and ground is greater than 250V.
If the fuse burns out during measurement, the Meter
may be damaged or the operator himself may be
hurt. Use proper terminals, function, and range for
the measurement. When the testing leads are
connected to the current terminals, do not parallel
them across any circuit.
The current measurement has 3 measurement positions
on the rotary switch: A Hz, mA Hz and Hz A.
The A Hz has a 400.0µA and 4000µA range, with
auto ranging; the mA Hz has a 40.00mA and 400.0mA
range, with auto ranging; Hz A position has a 4.000A
and 10.00A range, with auto ranging.
To measure current, do the following:
1. Turn off power to the circuit. Discharge all highvoltage capacitors.
2. Insert the red test lead into the AmA or 10A or
terminal and the black test lead into the COM
terminal.
3. Set the rotary switch to A Hz, mA Hz or Hz A.
Use the 10A terminal and Hz A measurement
position if the current value to be tested is an
unknown.
28
Measurement Operation (11)
4.
5.
6.
The Meter defaults to DC current measurement
mode. To toggle between DC and AC current
measurement function, press BLUE button.
AC current is displayed as an mean value (calibrated
against sine wave effective value).
Break the current path to be tested. Connect the
red test lead to the more positive side of the break
and the black test lead to the more negative side of
the break.
Turn on power to the circuit.
Note
l For safety sake, the measuring time for high current
should be 10 seconds for each measurement and
the interval time between 2 measurements should
be greater than 15 minutes.
l When current measurement has been completed,
29
The POWER button
This is a self-lock switch use to turn on or off the power
of the Meter.
The BLUE button
It uses for selecting the required measurement function
when there is more than one function at one position of
the rotary switch.
The REL mode applies to all measurement functions
except in frequency/duty cycle measurement, it subtracts
a stored value from the present value and displays the
relative value (∆) as the result.
The definition is as follows:
l Relative value (∆) = present value - stored value
For instance, if the stored value is 20.0V and the
present value is 22.0V, the reading would be 2.0V.
If a new measurement value is equal to the stored
value then display 0.0V.
To enter or exit REL mode:
l Use rotary switch to select the measurement function
before selecting REL∆. If measurement functions
change manually after REL∆ is selected, the Meter
exits the REL mode.
l Press REL∆ to enter REL mode, autoranging turns
off except under capacitance testing mode, and the
present measurement range is locked and display
the last measurement value as " 0 " as the stored
value.
l Press REL∆ again or turn the rotary switch to reset
the stored value and exit REL Mode.
Pressing HOLD H in REL mode makes the Meter stop
updating. Pressing HOLD H again to resume updating.
30
Warning
To avoid possibility of electric shock, do not use
Hold mode to determine if circuits are without power.
The Hold mode will not capture unstable or noisy
readings.
The Hold mode is applicable to all measurement functions.
l Press HOLD H to enter Hold mode; the Meter
beeps.
l Press HOLD H again or turn the rotary switch to
exit Hold mode; the Meter beeps.
l In Hold mode,
is displayed.
31
l
Maximum Voltage
between any Terminals: 1000V rms.
and Grounding
l
Fused Protection :0.5A, 250V fast type Glass
fuse, 5x20 mm.
for AmA Input
Terminal
l
Fused Protection :10A, 250V fast type Glass
for 10A Input Terminal fuse, 5x20 mm.
l
l
l
: Display:3999.
Maximum Display
Measurement Speed : Updates 3 times /second.
o
o
o
o
Temperature: Operating: 0 C~40 C (32 F ~104 F).
o
o
o
o
Storage : -10 C~50 C(14 F~122 F).
o
o
: 75% @ 0 C - 30 C;
Relative Humidity
o
o
50% @ 31 C- 40 C.
: 2000 m; Storage: 10000 m.
Altitude:Operating
: One piece of 9V NEDA1604
Battery Type
or 6F22 or 006P.
: Display .
Battery Deficiency
: Display .
Data Holding
: Display .
Negative reading
: Display .
Overloading
Equipped with full icons display.
Auto and manual range selectable.
Dimensions (HxWxL) : 177 x 85 x 40 mm.
: Approximate 300g (battery
Weight
included).
Safety/Compliances : IEC61010 CAT.III 1000V,
CAT.IV 600V overvoltage
standard.
Certification
UL & CUL
l
l
l
l
l
l
l
l
l
l
l
l
l
32
Accuracy Specifications (1)
Accuracy: (a% reading + b digits) guarantee for 1 year.
o
o
Operating temperature: 23 C 5 C.
Relative humidity: & lt; 75%.
o
Temperature coefficient: 0.1 x (specified accuracy) / 1 C.
A. AC Voltage
Range Resolution
4V
1mV
40V
10mV
400V
100mV
750V
1V
Overload
Protection
Accuracy
(1%+5)
1000V DC
750V AC rms
continuous.
(1.2%+5)
Remarks:
l Input impedance 10M .
l Displays effective value of sine wave (mean value
response).
l Frequency response 40Hz ~ 400Hz.
B. DC Voltage
Range
Resolution
400mV
0.1mV
4V
1mV
40V
10mV
400V
100mV
1000V
1V
Overload
Protection
Accuracy
(0.8%+3)
1000V DC
750V AC rms
continuous.
(0.8%+1)
(1%+3)
Remark: Input impedance 10M
33
Accuracy Specifications (2)
C. Continuity, Diodes & Resistance Test
Range
Continuity Test
(400.0 )
Diode
400
4k
40k
400k
4M
40M
Resolution Accuracy
0.1
1mV
0.1
1
10
100
1k
10k
Approximate
100
N/A
(1.2%+2)
Overload
Protection
600Vp
(1%+2)
(1.2%+2)
(1.5%+2)
Remarks:
l Continuity Test (400.0 ) Range:
Buzzer beeps continuously.
Open circuit voltage approximate 0.45V.
l Diode Range:
Open circuit voltage approximate 1.48V.
Displays approximate forward voltage drop
reading 0.5V~0.8V.
l 400 ~ 40M Range:
34
Accuracy Specifications (3)
D. Capacitance
Range Resolution
40nF
400nF
4 F
40 F
100 F
Accuracy
10pF
100pF
1nF
10nF
100nF
Overload
Protection
(3%+5)
600Vp
(4%+5)
E. Frequency & Duty Cycle
Range
10Hz~
10MHz
0.1%~
99.9%
(Duty cycle)
Resolution
N/A
Accuracy
Overload
Protection
(0.1%+3)
600Vp
0.01%
N/A
Remarks:
l 10Hz~10MHz Range:
Input sensitivity as follows:
1MHz: 300mV rms;
& gt; 1MHz: 600mV rms.
Maximum input amplitude: 30V rms.
l 0.1%~99.9% Range:
Reading is only for reference purpose.
35
Accuracy Specifications (4)
F.
DC Current
Overload
Protection
Range Resolution Accuracy
400 A
4000 A
40mA
400mA
0.1 A
1 A
0.01mA
0.1mA
4A
0.001A
10A
0.01A
(1%+2)
(1.2%+3)
0.5A, 250V,
fast type Glass
fuse, 5x20mm
10A, 250V,
(1.5%+5) fast type Glass
fuse, 5x20mm
Remark:
l 4A & 10A Range:
For continuous measurement 10 seconds and
interval time between 2 measurements greater than
15 minutes.
G. AC Current
Range
Resolutio
400 A
4000 A
40mA
400mA
0.1 A
1 A
0.01mA
0.1mA
4A
Overload
Protection
(1.5%+5) 0.5A, 250V,
fast type Glass
(2%+5) fuse, 5x20mm
0.001A
10A
Accuracy
0.01A
10A, 250V,
(2.5%+5) fast type Glass
fuse, 5x20mm
Remarks:
l Displays effective value of sine wave (mean value response).
l 4A & 10A Range:
15 minutes.
36
Maintenance (1)
This section provides basic maintenance information
including battery and fuse replacement instruction.
Warning
Do not attempt to repair or service your Meter unless
you are qualified to do so and have the relevant
calibration, performance test, and service information.
To avoid electrical shock or damage to the Meter, do
not get water inside the case.
A. General Service
l
l
l
l
Periodically wipe the case with a damp cloth and
mild detergent. Do not use abrasives or solvents.
To clean the terminals with cotton bar with detergent,
as dirt or moisture in the terminals can affect readings.
Press the POWER to turn off the Meter when it is
not in use and take out the battery when not using
for a long time.
Do not store the Meter in a place of humidity, high
temperature and strong magnetic field.
B. Testing the Fuses
Warning
To avoid electrical shock or personal injury, remove
the test leads and any input signals before replacing
the battery or fuse.
To prevent damage or injury, install ONLY replacement
fuses with identical amperage, voltage, and speed
ratings.
37
Maintenance (2)
To test the fuse:
1. Set the rotary switch to
and select
by
pressing BLUE button.
2. Plug a test lead into the terminal HzV and touch
the probe tip to the 10A or AmA terminal.
l If the Meter beeps, the fuse is good.
l If the display shows , replace the fuse and
test again.
l If the display shows any other value, have the
Meter serviced and contact your dealer
immediately.
If the Meter does not work while the fuse is all right, send
it to your dealer for repair.
C. Replacing the Battery (see figure 11)
Screw
(figure 11)
Warning
To avoid false readings, which could lead to possible
electric shock or personal injury, replace the battery
as soon as the battery indicator " " appears.
To replace the battery:
1.
Press the POWER to turn the Meter off and remove
all connections from the terminals.
38
Maintenance (3)
2.
3.
4.
5.
Remove the screw from the battery compartment,
and separate the battery compartment from the case
bottom.
Remove the battery from the battery compartment.
Replace the battery with a new 9V battery
(NEDA1604, 6F22 or 006P).
Rejoin the case bottom and battery compartment,
and reinstall the screw.
D. Replacing the Fuses (see figure 12)
Rubber foot
Screw
(figure 12)
Warning
To avoid electrical shock or arc blast, or personal
injury or damage to the Meter, use specified fuses
ONLY in accordance with the following procedure.
To replace the Meter's fuse:
1.
2.
3.
4.
bottom.
Remove the 2 rubber feet and 2 screws from the
case bottom, and separate the case top from the
case bottom.
Remove the fuse by gently prying one end loose,
and then take out the fuse from its bracket.
39
Maintenance (4)
5.
6.
7.
Install ONLY replacement fuses with the identical
type and specification as follows and make sure the
fuse is fixed firmly in the bracket.
Fuse 1:0.5A, 250V, fast type Glass fuse,
5x20mm.
Fuse 2: 10A, 250V, fast type Glass fuse,
5x20 mm.
Rejoin the battery compartment and the case top,
Rejoin the case bottom and case top, and reinstall
the 2 screws and 2 rubber feet.
Replacement of the fuses is seldom required. Burning
of a fuse always results from improper operation.
40
RS232C Serial Port (1)
A. RS232C Port Cable
The
Meter
D-sub
9 Pin
Male
2
Computer
D-sub
D-sub
9 Pin 25 Pin
Female Female
2
3
Pin
Name
Remark
RX
Receiving
Data
Transmitting
Data
Data Terminal
Ready
3
3
2
TD
4
4
20
DTR
5
6
5
6
7
6
GND
DSR
7
7
4
RTS
8
8
5
CTS
Grounding
Data Set
Ready
Request To
Send
Clear To Send
B. Setting of RS232C Serial Ports
Default of RS232C serial port for communication is set
as:
Baud Rate 2400
Start bit
1 (always 0)
Stop bit
1 (always 1)
Data bits 8
Parity
1 (Odd)
41
RS232C Serial Port (2)
C. System Requirements for Installing the UT60A
Interface Program
To use UT60A Interface Program, you need the following
hardware and software:
l An IBM PC or equivalent computer with 80486 or
higher processor and 640 x 480 pixel or better
monitor.
l Microsoft Windows 95 or above.
l At least 8MB of RAM.
l At least 8MB free space in hard drive.
l Can access to a local or a network CD-ROM.
l A free serial port.
l A mouse or other pointing device supported by
Windows.
Please refer to the included " Installation Guide &
Computer Interface Software " for installing and
operating instructions of the UT60A Interface Program.
~ END ~
This operating manual is subject to change without notice.
42
43
Copyright 2001 Uni-Trend International Limited.
All rights reserved.
Manufacturer: UNI-TREND TECHNOLOGY(DONG GUAN)LIMITED
Address: Dong Fang Da Dao, Bei Shan Dong Fang Industrial
Development District, Hu Men Town, Dong Guan City,
Guang Dong Province, China
Headquarters: Uni-Trend International Limited
Address: Rm901, 9/F, Nanyang Plaza 57 Hung To Road
Kwun Tong Kowloon, Hong Kong
Tel: (852) 2950 9168
Fax: (852) 2950 9303
Email: info@uni-trend.com
http://www.uni-trend.com
44
