Service Manual
TDS 200-Series
Digital Real-Time Oscilloscope
071-0492-02
This document supports firmware version 1.00
and above.
Warning
The servicing instructions are for use by
qualified personnel only. To avoid personal
injury, do not perform any servicing unless you
are qualified to do so. Refer to all safety
summaries prior to performing service.
www.tektronix.com
Copyright © Tektronix, Inc. All rights reserved.
Tektronix products are covered by U.S. and foreign patents, issued and pending. Information in this publication supercedes
that in all previously published material. Specifications and price change privileges reserved.
Tektronix, Inc., P.O. Box 500, Beaverton, OR 97077
TEKTRONIX and TEK are registered trademarks of Tektronix, Inc.
WARRANTY
Tektronix warrants that the products that it manufactures and sells will be free from defects in materials and workmanship for
a period of three (3) years from the date of purchase from an authorized Tektronix distributor. If any such product proves
defective during this warranty period, Tektronix, at its option, either will repair the defective product without charge for parts
and labor, or will provide a replacement in exchange for the defective product. Batteries are excluded from this warranty.
In order to obtain service under this warranty, Customer must notify Tektronix of the defect before the expiration of the
warranty period and make suitable arrangements for the performance of service. Customer shall be responsible for packaging
and shipping the defective product to the service center designated by Tektronix, shipping charges prepaid, and with a copy of
customer proof of purchase. Tektronix shall pay for the return of the product to Customer if the shipment is to a location
within the country in which the Tektronix service center is located. Customer shall be responsible for paying all shipping
charges, duties, taxes, and any other charges for products returned to any other locations.
This warranty shall not apply to any defect, failure or damage caused by improper use or improper or inadequate maintenance
and care. Tektronix shall not be obligated to furnish service under this warranty a) to repair damage resulting from attempts
by personnel other than Tektronix representatives to install, repair or service the product; b) to repair damage resulting from
improper use or connection to incompatible equipment; c) to repair any damage or malfunction caused by the use of
non-Tektronix supplies; or d) to service a product that has been modified or integrated with other products when the effect of
such modification or integration increases the time or difficulty of servicing the product.
THIS WARRANTY IS GIVEN BY TEKTRONIX WITH RESPECT TO THE LISTED PRODUCTS IN LIEU OF
ANY OTHER WARRANTIES, EXPRESS OR IMPLIED. TEKTRONIX AND ITS VENDORS DISCLAIM ANY
IMPLIED WARRANTIES OF MERCHANTABILITY OR FITNESS FOR A PARTICULAR PURPOSE.
TEKTRONIX’ RESPONSIBILITY TO REPAIR OR REPLACE DEFECTIVE PRODUCTS IS THE SOLE AND
EXCLUSIVE REMEDY PROVIDED TO THE CUSTOMER FOR BREACH OF THIS WARRANTY. TEKTRONIX
AND ITS VENDORS WILL NOT BE LIABLE FOR ANY INDIRECT, SPECIAL, INCIDENTAL, OR
CONSEQUENTIAL DAMAGES IRRESPECTIVE OF WHETHER TEKTRONIX OR THE VENDOR HAS
ADVANCE NOTICE OF THE POSSIBILITY OF SUCH DAMAGES.
Table of Contents
General Safety Summary . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Service Safety Summary . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Preface . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
v
vii
ix
Related Manuals . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Contacting Tektronix . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Product End-of-Life Handling . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
ix
x
x
Specifications
Specifications . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
1–1
Operating Information . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
2–1
General Features . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Installation . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Extension Modules . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Functional Check . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Probe Compensation . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Self Calibration . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Probe Safety . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Probe Attenuation Setting . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Factory Setup . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
2–2
2–3
2–4
2–5
2–6
2–7
2–7
2–7
2–8
Theory of Operation . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
3–1
Main Board . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Power Supply . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Display Module . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Front Panel . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
3–1
3–2
3–2
3–2
3–2
Theory of Operation
Performance Verification
Performance Verification . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
4–1
Test Record . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Performance Verification Procedures . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
4–2
4–3
Adjustment Procedures
Adjustment Procedures . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
5–1
Required Equipment . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Adjustment Procedure . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
5–1
5–2
TDS 200 Series Digital Oscilloscope Service Manual
i
Table of Contents
Maintenance
Maintenance . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
6–1
Preparation . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Preventing ESD . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Inspection and Cleaning . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Removal and Installation Procedures . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Troubleshooting . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Repackaging Instructions . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
6–1
6–1
6–2
6–5
6–38
6–48
Options . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
7–1
Electrical Parts List . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
8–1
Diagrams . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
9–1
Replaceable Parts . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
10–1
Parts Ordering Information . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Using the Replaceable Parts List . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
10–1
10–2
Options
Diagrams
Replaceable Parts
ii
Table of Contents
List of Figures
Figure 2–1: Routing the power cord and security cable . . . . . . . . . . .
Figure 2–2: Installing an extension module . . . . . . . . . . . . . . . . . . . . .
2–3
2–4
Figure 3–1: Module-level block diagram (two channel) . . . . . . . . . . .
Figure 3–2: Module-level block diagram (four channel) . . . . . . . . . . .
3–3
3–4
Figure 5–1: Service menu enable button . . . . . . . . . . . . . . . . . . . . . . . .
Figure 5–2: Adjustment setups . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
5–2
5–3
Figure 6–1: Removing the rear feet . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Figure 6–2: Installing the rear feet . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Figure 6–3: Installing a front-case label . . . . . . . . . . . . . . . . . . . . . . . .
Figure 6–4: Installing a new rear-case label . . . . . . . . . . . . . . . . . . . . .
Figure 6–5: Removing the handle . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Figure 6–6: Installing the handle . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Figure 6–7: Removing and installing the rear cover . . . . . . . . . . . . . .
Figure 6–8: Removing and installing the front feet . . . . . . . . . . . . . . .
Figure 6–9: Removing and installing the EMI clips . . . . . . . . . . . . . . .
Figure 6–10: Removing the flip stand . . . . . . . . . . . . . . . . . . . . . . . . . .
Figure 6–11: Installing the flip stand . . . . . . . . . . . . . . . . . . . . . . . . . . .
Figure 6–12: Line fuse location . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Figure 6–13: Removing the power supply module . . . . . . . . . . . . . . . .
Figure 6–14: Installing the power supply module . . . . . . . . . . . . . . . .
Figure 6–15: Removing and installing the internal assembly . . . . . . .
Figure 6–16: Installing the copper mesh grounding tube . . . . . . . . . .
Figure 6–17: Main board removal . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Figure 6–18: Main board installation . . . . . . . . . . . . . . . . . . . . . . . . . .
Figure 6–19: Removing the display module . . . . . . . . . . . . . . . . . . . . .
Figure 6–20: Installing the display module . . . . . . . . . . . . . . . . . . . . . .
Figure 6–21: Removing the front panel module . . . . . . . . . . . . . . . . . .
Figure 6–22: Installing the front panel module . . . . . . . . . . . . . . . . . .
Figure 6–23: Removing and installing the keypad . . . . . . . . . . . . . . . .
Figure 6–24: Removing the display shield . . . . . . . . . . . . . . . . . . . . . .
Figure 6–25: Installing the display shield . . . . . . . . . . . . . . . . . . . . . . .
Figure 6–26: Measuring the backlight voltage . . . . . . . . . . . . . . . . . . .
6–8
6–9
6–10
6–11
6–12
6–13
6–15
6–16
6–17
6–18
6–19
6–20
6–21
6–22
6–24
6–25
6–26
6–27
6–28
6–30
6–31
6–32
6–33
6–35
6–36
6–41
Figure 10–1: Exploded diagram . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
10–7
iii
Table of Contents
List of Tables
Table 1–1: Specifications . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Table 1–2: General specifications . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Table 2–1: Factory setup settings . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
2–8
Table 5–1: Required equipment . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Table 5–2: TDS 210 and TDS 220 Adjustment steps . . . . . . . . . . . . . .
Table 5–3: TDS 224 Adjustment steps . . . . . . . . . . . . . . . . . . . . . . . . .
5–1
5–5
5–7
Table 6–1: Internal inspection check list . . . . . . . . . . . . . . . . . . . . . . .
Table 6–2: List of procedures . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Table 6–3: List of error codes . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
6–3
6–7
6–46
Table 10–1: Parts list column descriptions . . . . . . . . . . . . . . . . . . . . . .
Table 10–2: Manufacturers cross index . . . . . . . . . . . . . . . . . . . . . . . .
Table 10–3: Replaceable parts list . . . . . . . . . . . . . . . . . . . . . . . . . . . . .
Table 10–4: Replaceable standard accessories . . . . . . . . . . . . . . . . . . .
Table 10–5: Replaceable optional accessories . . . . . . . . . . . . . . . . . . .
iv
1–1
1–5
10–2
10–3
10–3
10–8
10–8
Review the following safety precautions to avoid injury and prevent damage to
this product or any products connected to it.
Only qualified personnel should perform service procedures.
Injury Precautions
Use Proper Power Cord. To avoid fire hazard, use only the power cord specified
for this product.
Avoid Electric Overload. To avoid electric shock or fire hazard, do not apply a
voltage to a terminal that is outside the range specified for that terminal.
Avoid Overvoltage. To avoid electric shock or fire hazard, do not apply potential
to any terminal, including the common terminal, that varies from ground by more
than the maximum rating for that terminal.
Avoid Electric Shock. To avoid injury or loss of life, do not connect or disconnect
probes or test leads while they are connected to a voltage source.
Ground the Product. This product is grounded through the grounding conductor
of the power cord. To avoid electric shock, the grounding conductor must be
connected to earth ground. Before making connections to the input or output
terminals of the product, ensure that the product is properly grounded.
Connect the Probe Properly. The probe ground lead is at ground potential. Do not
connect the ground lead to an elevated voltage.
Do Not Operate Without Covers. To avoid electric shock or fire hazard, do not
operate this product with covers or panels removed.
Use Proper Fuse. To avoid fire hazard, use only the fuse type and rating specified
for this product.
Do Not Operate in Wet/Damp Conditions. To avoid electric shock, do not operate
this product in wet or damp conditions.
Do Not Operate in an Explosive Atmosphere. To avoid injury or fire hazard, do not
operate this product in an explosive atmosphere.
Product Damage
Precautions
Use Proper Power Source. Do not operate this product from a power source that
applies more than the voltage specified.
Provide Proper Ventilation. To prevent product overheating, provide proper
ventilation.
Do Not Operate With Suspected Failures. If you suspect there is damage to this
product, have it inspected by qualified service personnel.
v
Symbols and Terms
Terms in this Manual. These terms may appear in this manual:
WARNING. Warning statements identify conditions or practices that could result
in injury or loss of life.
CAUTION. Caution statements identify conditions or practices that could result in
damage to this product or other property.
Terms on the Product. These terms may appear on the product:
DANGER indicates an injury hazard immediately accessible as you read the
marking.
WARNING indicates an injury hazard not immediately accessible as you read the
marking.
CAUTION indicates a hazard to property including the product.
Symbols on the Product. The following symbols may appear on the product:
DANGER
High Voltage
Certifications and
Compliances
vi
Protective Ground
(Earth) Terminal
ATTENTION
Refer to Manual
Double
Insulated
Refer to the specifications section for a listing of certifications and compliances
that apply to this product.
Only qualified personnel should perform service procedures. Read this Service
Safety Summary and the General Safety Summary before performing any service
procedures.
Do Not Service Alone. Do not perform internal service or adjustments of this
product unless another person capable of rendering first aid and resuscitation is
present.
Disconnect Power. To avoid electric shock, disconnect the main power by means
of the power cord or, if provided, the power switch.
Use Care When Servicing With Power On. Dangerous voltages or currents may
exist in this product. Disconnect power, remove battery (if applicable), and
disconnect test leads before removing protective panels, soldering, or replacing
components.
To avoid electric shock, do not touch exposed connections.
vii
viii
Preface
The service manual for the TDS 200-Series Digital Real-Time Oscilloscope
provides information to troubleshoot and repair the instrument to the module
level.
Some TDS models have two input channels and an external trigger while other
models have four input channels. Most illustrations in this manual show a two
channel model. Some parts for the four channel model are slightly different
because of the additional channels.
Related Manuals
Additional documentation for the instrument is contained in the related manuals
listed below.
Language
User manual
part number
Extension module
instructions part
number
Programmer manual
part number
English
071-0398-XX
071-0409-XX
071-0493-XX
French
071-0400-XX*
071-0483-XX
German
071-0402-XX*
071-0485-XX
Italian
071-0401-XX*
071-0484-XX
Spanish
071-0399-XX*
071-0482-XX
Portuguese
071-0403-XX*
071-0486-XX
Japanese
071-0405-XX*
071-0488-XX
Korean
071-0408-XX*
071-0491-XX
Simplified Chinese
071-0406-XX*
071-0489-XX
Traditional Chinese
071-0407-XX*
071-0490-XX
Russian
071-0404-XX
071-0487-XX
*These manuals contain a language overlay for the front-panel controls.
ix
Preface
Phone
1-800-833-9200*
Address
Tektronix, Inc.
Department or name (if known)
14200 SW Karl Braun Drive
P.O. Box 500
Beaverton, OR 97077
USA
Web site
www.tektronix.com
Sales support
1-800-833-9200, select option 1*
Service support
1-800-833-9200, select option 2*
Technical support
Email: techsupport@tektronix.com
1-800-833-9200, select option 3*
1-503-627-2400
6:00 a.m. – 5:00 p.m. Pacific time
This phone number is toll free in North America. After office hours, please leave a
voice mail message.
Outside North America, contact a Tektronix sales office or distributor; see the
Tektronix web site for a list of offices.
Components that Contain Mercury. The cold cathode fluorescent tube located in
the liquid crystal display backlight contains trace elements of mercury. When you
are ready to reclaim the instrument, you must properly transfer it according to
local regulations concerning mercury-containing equipment or ship the instrument to the Tektronix Recycling Operations (RAMS). You can contact
Tektronix for the RAMS shipping address and instructions.
x
Specifications
All specifications apply to the TDS 200-Series Digital Real-Time Oscilloscope
with a P2100 probe with the Attenuation switch set to 10X unless noted
otherwise. To meet specifications, two conditions must first be met:
H
The instrument must have been operating continuously for twenty minutes
within the specified operating temperature.
H
You must perform the Self Cal operation, accessible through the utility menu,
if the operating temperature changes by more than 5° C.
All specifications are guaranteed unless noted “typical.” Specifications that are
marked with the n symbol are checked in the chapter Performance Verification.
Acquisition
Acquisition Modes
Sample, Peak detect, and Average
Acquisition Rate,
typical
Up to 180 waveforms per second, per channel (Sample acquisition mode, no measurements)
Single Sequence
Acquisition Mode
Acquisition Stops After
Sample, Peak Detect
Single acquisition, all channels simultaneously
Average
N acquisitions, all channels simultaneously, N is
selectable from 4, 16, 64, and 128
Inputs
Input Coupling
DC, AC, or GND
Input Impedance,
DC Coupled, all channela
1 MW ±2% in parallel with 20 pF ±3 pF
Input Impedance,
DC Coupled,
EXT TRIG only
TDS 210 (B099188 to B119999 & C021679 to C029999)
TDS 220 (B065810 to B079999 & C021127 to C029999)
All other TDS 210, TDS 220, and TDS 224
1.2 MW ±5% in parallel with 20 pF ±5 pF
1 MW ±5% in parallel with 20 pF ±5 pF
P2100 Probe Attenuation
1X, 10X
Probe Attenuation Factors
1X, 10X, 100X, 1000X
Maximum Voltage Between Overvoltage Category
Signal and Common at
CAT I and CAT II
input BNC
CAT III
Maximum Voltage
300 VRMS (420 V peak, duty factor & lt; 50%, pulse
width & lt; 100 msec.)
150 VRMS
For steady-state sinusoidal waveforms, derate at 20 dB/decade above 100 kHz to 13 Vpk at 3 MHz* and
above. Also, refer to Overvoltage Category description on page 1–7.
Bandwidth is not valid for the P2100 probe when the switch is set to 1X.
1–1
Specifications
Table 1–1: Specifications (cont.)
Inputs
Probe Tip and ground
CAT I and CAT II
using P2100 connected to
input BNC
CAT III
Maximum Voltage
300 VRMS (500 V peak, duty factor & lt; 35%, pulse
100 VRMS
Derate at 20 dB/decade above 900 kHz to 13 VRMS at 27 MHz* and above. Also, refer to Overvoltage
Time delay between
channels, typical
150 ps
Channel-to-Channel
Common Mode Rejection,
typical
TDS 210
TDS 220 and TDS 224
100:1 at 60 Hz
20:1 at 30 MHz*
100:1 at 60 Hz
20:1 at 50 MHz*
Measured on MATH Ch1 – Ch2 waveform, with test signal applied between signal and common of both
channels, and with the same VOLTS/DIV and coupling settings on each channel. Also measured on MATH
Ch3 – Ch4 waveform for the TDS 224.
Channel-to-Channel Cross- TDS 210
talk
w100:1 at 30 MHz*
TDS 220 and TDS 224
w100:1 at 50 MHz*
Measured on one channel, with 8 division test signal applied between signal and common of the other
channel, the same VOLTS/DIV and coupling settings on each channel, and 50 W terminators on each
channel.
Vertical
Digitizers
8 bit resolution (except when set to 2 mV/div), each channel sampled simultaneously
VOLTS/DIV Range
2 mV/div to 5 V/div at input BNC
(Full bandwidth at u5 mV/div to 5 V/div, 20 MHz at 2 mV/dif to 5 mV/div, except in Peak Detect mode full
bandwidth at u10 mV/div to 5 V/div [20 MHz at 2 mV/div to 10 mV/div])
Position Range
2 mV/div to 200 mV/div, ±2 V
& gt; 200 mV/div to 5 V/div, ±50 V
n Analog Bandwidth in
TDS 210
TDS 220 and TDS 224
Sample and Average modes
60 MHz* (when vertical scale set to & gt; 5 mV/div)
at BNC or with P2100 probe,
DC Coupled
20 MHz* (when vertical scale set to ≤ 5 mV/div)
Analog Bandwidth in Peak
Detect mode (5 s/div to
5m
ms/div), typical
100 MHz* (when vertical scale set to & gt; 5 mV/div)
TDS 210
TDS 220 and TDS 224
50 MHz* (when vertical scale set to & gt; 10 mV/div)
75 MHz* (when vertical scale set to & gt; 10 mV/div)
20 MHz* (when vertical scale set to ≤ 10 mV/div)
Selectable Analog Bandwidth Limit, typical
20 MHz*
Lower Frequency Limit,
AC Coupled
≤10 Hz at BNC
1–2
≤1 Hz when using a 10X passive probe
Specifications
Vertical
Rise Time at BNC, typical
TDS 210
TDS 220 and TDS 224
& lt; 5.8 ns
& lt; 3.5 ns
Peak Detect Response
Captures 50% or greater amplitude of pulses ≥10 ns wide (5 s/div to 5 ms/div) in the center 6 vertical
divisions. The oscilloscope reverts to Sample mode when the sec/div (Horizontal scale) setting is from
2.5 ms/div to 5 ns/div. The Sample mode can still capture 10 ns glitches.
DC Gain Accuracy
2 mV/div to 5 mV/div, +4% for Sample or Average acquisition mode
10 mV/div and larger, +3% for Sample or Average acquisition mode
Accuracy
n DC Measurement Accu- Measurement Type
racy, Average Acquisition
Average of ≥16 waveforms with vertical position at zero ±(4% × reading + 0.1 div + 1 mV) and either
Mode
2 mV/div or 5 mV/div is selected.
±(3% × reading + 0.1 div + 1 mV) and 10 mV/div
or greater is selected.
Average of ≥16 waveforms with vertical position not at
zero
±[3% × (reading + vertical position) + 1% of
vertical position + 0.2 div]
Add 2 mV for settings from 2 mV/div to 200 mV/
div. Add 50 mV for settings from u200 mV/div to
5 V/div.
Delta Volts Measurement
Delta volts between any two averages of ≥16
Accuracy, Average Acquisi- waveforms acquired under same setup and ambient
tion Mode
conditions
3% of reading + 0.05 div
P2100 passive probe
10X position
1X position
Bandwidth
DC to 100 MHz
DC to 7 MHz
Attenuation ratio
10:1
1:1
Compensation Range
18 pf-35 pf
All oscilloscopes with 1 M W input
Input resistance
10 M W ± 2%
1 M W ± 2%
Input capacitance
14.5 pf-17.5 pf
80 pf-110 pf
Maximum input voltage
300 VRMS CAT I or 300 V DC CAT I
300 VRMS CAT II or 300 V DC CAT II
100 VRMS CAT III or 100 V DC CAT III
150 VRMS CAT I or 150 V DC CAT I
150 VRMS CAT II or 150 V DC CAT II
Horizontal
Sample Rate Range
50 S/s to 1 GS/s
Waveform Interpolation
(sin x)/x
Record Length
2500 samples for each channel
SEC/DIV Range
5 ns/div to 5 s/div, in a 1, 2.5, 5 sequence
n Sample Rate and Delay ±100 ppm over any ≥1 ms time interval
Time Accuracy
1–3
Specifications
Horizontal
Delta Time Measurement
Accuracy (Full Bandwidth)
Conditions
Accuracy
Single-shot, Sample mode
±(1 sample interval + 100 ppm × reading +
0.6 ns)
& gt; 16 averages
0.4 ns)
Sample interval = s/div ÷ 250
5 ns/div to 10 ns/div
25 ns/div to 100 ms/div
(–4 div × s/div) to 20 ms
Position Range
250 ms/div to 5 s/div
(–4 div × s/div) to 50 ms (–4 div × s/div) to 50 s
Trigger
n Trigger Sensitivity, Edge Coupling
Trigger Type
DC
Sensitivity
CH 1, CH 2,
CH 3 & CH 4
1 div from DC to 10 MHz*,
1.5 div from 10 MHz* to Full
TDS 210 and TDS 220
EXT
EXT/5
Trigger Type, typical
100 mV from DC to 10 MHz*,
150 mV from 10 MHz* to Full
500 mV from DC to 10 MHz*,
750 mV from 10 MHz* to Full
Sensitivity
AC
Same as DC at 50 MHz and above
NOISE REJ
Reduces DC-coupled trigger sensitivity by 2 times for & gt; 10 mV/div to 5 V/div
HF REJ
Same as DC-coupled limit from DC to 7 kHz, attenuates signals above
80 kHz
LF REJ
Same as the DC-coupled limits for frequencies above 300 kHz, attenuates
signals below 300 kHz
Source
Range
Internal
Trigger Level Range
Coupling
±8 divisions from center of screen
TDS 210 and TDS 220
EXT
EXT/5
Trigger Level Accuracy,
typical
±1.6 V
±8 V
Accuracies are for signals having rise and fall times ≥20 ns
Source
Accuracy
Internal
±0.2 div
volts/div within ±4 divisions from center screen
TDS 210 and TDS 220
EXT
EXT/5
1–4
±(6% of setting + 40 mV)
±(6% of setting + 200 mV)
Specifications
Trigger
SET LEVEL TO 50%,
typical
Operates with input signals ≥50 Hz
Default Settings, Video
Trigger
Trigger mode is Auto and Coupling is AC
Sensitivity, Video
Composite video signal
Source
Range
Internal
Pk-pk amplitude of 2 divisions
TDS 210 and TDS 220
EXT
400 mV
EXT/5
2V
Signal Formats and Field
Rates, Video Trigger Type
Supports NTSC, PAL, and SECAM broadcast systems for any field or any line
Holdoff Range
500 ns to 10 s
Measurements
Cursors
Voltage difference between cursors (DV)
Time difference between cursors (DT)
Reciprocal of DT in Hertz (1/DT)
Automated Measurements
Cycle RMS, Mean, Pk – Pk, Period, Frequency
Display
Display Type
5.9 in (152 mm) diagonal liquid crystal
Display Resolution
320 horizontal by 240 vertical pixels
Display Contrast
Adjustable, temperature compensated
Backlight Intensity, typical
60 cd/m2
Probe compensator output
Output Voltage, typical
5 V into ≥1 MW load
Frequency, typical
1 kHz
Power Source
Source Voltage
100 – 120 VACRMS " 10% from 45 Hz through 440 Hz, CAT II
120 – 250 VACRMS " 10% from 45 Hz through 66 Hz, CAT II
Power Consumption
TDS 224
Less than 20 W
Fuse
TDS 210 and TDS 220
Less than 25 W
1 A, T rating, 250 V
1–5
Specifications
Table 1–2: General specifications (cont.)
Environmental
Temperature
Operating
0° C to +50° C
Nonoperating
–20° C to +60° C
Cooling Method
Convection
Humidity
+40° C or below
≤90% relative humidity
+41° C to +50° C
≤60% relative humidity
Operating
3,000 m
Nonoperating
15,000 m
Operating
0.31 gRMS from 5 Hz to 500 Hz, 10 minutes on each
axis
Nonoperating
2.46 gRMS from 5 Hz to 500 Hz, 10 minutes on each
axis
Operating
50 g, 11 ms, half sine
Altitude
Random Vibration
Mechanical Shock
Mechanical
Size
Height
Width
304.8 mm (12 in.)
Depth
Weight (approximate)
151.4 mm (5.96 in.)
120.7 mm (4.75 in.)
When packaged for domestic shipment
3.6 kg (8.0 lbs)
Certifications and compliances
EC Declaration of
Conformity
Meets intent of Directive 89/336/EEC for Electromagnetic Compatibility*; meets the requirements of FCC
Code of Federal Regulations, 47 CFR, Part 15, Subpart B, Class A:
(TDS 210, TDS 220 &
TDS 224)
CISPR 11 (Class A): Electromagnetic radiated and conducted emissions
EN 50082-1 European Community Requirements:
IEC 61000-4-2 electrostatic discharge
IEC 61000-4-3 radiated susceptibility**
IEC 61000-4-4 power line fast transients
IEC 61000-4-5 power line surge immunity
IEC 61000-4-6 conducted immunity to RF fields
IEC 61000-4-11 voltage dips, interrupts and variations
AS/NZS 2064, Australian emissions standard for Industrial, Scientific, and Medical Equipment
(TDS 210, TDS 220,
TDS 224, P2100)
(P2100)
Low Voltage Directive 73/23/EEC as amended by 93/68/EED:
EN61010-1/A2:1995
Safety requirements for electrical equipment for measurement control, and
laboratory use
EN 61010-2-031:1994
Particular requirements for hand-held probe assemblies for electrical
measurement and test
Instrument tested with shielded cables.
** Meets Performance Criterion A with allowed degradation of up to 5 major divisions of trace noise increase from 80 MHz to
200 MHz and up to 2 divisions of trace noise increase over the range of 200 MHz to 1000 MHz under a 3 V/m RF field. Ambient RF
fields may induce triggering when trigger threshold is offset less than 2.5 major divisions from ground reference.
1–6
Specifications
Certifications
(TDS 210, TDS 220,
TDS 224)
(P2100)
CAN/CSA C22.2 No. 1010.1-92
UL3111-1, First Edition
IEC61010-1/A2
IEC61010-2-031
CSA Certified Power Cords CSA Certification includes the products and power cords appropriate for use in the North America power
network. All other power cords supplied are approved for the country of use.
Pollution Degree 2
Do not operate in environments where conductive pollutants may be present.
Category:
Examples of Products in this Category:
CAT III
Distribution-level mains, fixed installation
CAT II
Local-level mains, appliances, portable equipment
CAT I
Signal levels in special equipment or parts of equipment, telecommunications,
electronics
Adjustment interval
The recommended adjustment interval is one year
1–7
Specifications
1–8
TDS 200-Series Digital Oscilloscopes are two or four channel oscilloscopes in
small, lightweight, benchtop packages that you can use to take ground-referenced
measurements.
In addition to the list of general features, this chapter covers the following topics:
H
How to install your product
H
How to add extended functions
H
How to perform a brief functional check
H
How to compensate probes
H
How to use the self calibration routine
H
How to match your probe attenuation factor
For more detailed information about instrument operation, refer to your user
manual.
2–1
General Features
H
H
1 GS/s sample rate and 2,500 point record length for each channel
H
Cursors with readout
H
Five automated measurements
H
High-resolution, high-contrast LCD display with temperature compensation
and replaceable backlight
H
Setup and waveform storage
H
Autoset for quick setup
H
Waveform averaging and peak detection
H
Digital real-time oscilloscope (at least ten-times over sampling)
H
Dual time base
H
Video trigger capability
H
RS-232, GPIB, and Centronics communication ports easily added with
optional extension modules
H
Variable persistence display
H
2–2
100 MHz (TDS 220 and TDS 224) or 60 MHz (TDS 210) bandwidth with
selectable 20 MHz bandwidth limit
User interface available in ten user selectable languages
Installation
Power Cord
Use only power cords designed for your oscilloscope. Use a power source that
delivers 85 to 275 VACRMS, 47 to 63 Hz. Refer to page 10–9 for a list of
available power cords.
Use the power cord notch to help route the cord to the rear of the instrument and
avoid inadvertently disconnecting the power source.
Power cord
notch
Securing cable
Security Loop
Use the built-in cable channels to secure both your instrument and extension
module to your location.
2–3
Extension Modules
You can increase the feature set of your oscilloscope by inserting an extension
module. Refer to page 10–9 for information about the available modules.
Figure 2–2 shows the proper way to install a module without bending the
connecting pins.
CAUTION. Electrostatic discharge (ESD) can damage components in the
extension module and the oscilloscope. Do not operate your instrument with the
extension module connector exposed.
1
2
2–4
Functional Check
Perform this quick functional check to verify that your instrument is operating
correctly.
1. Turn on the instrument.
ON/OFF
button
Wait until the display shows that all self tests passed. Push the SAVE/RECALL
button, select Setups in the top menu box and push the Recall Factory menu
box. The default Probe menu attenuation setting is 10X.
PASSED
2. Set the switch to 10X on the P2100 probe and connect the oscilloscope probe to
PROBE COMP
channel 1. To do this, align the slot in the probe connector with the key on the
CH 1 BNC, push to connect, and twist to the right to lock the probe in place.
Attach the probe tip and reference lead to the PROBE COMP connectors.
CH 1
3. Push the AUTOSET button. Within a few seconds, you should see a square
wave in the display (approximately 5 V at 1 kHz peak-to-peak).
Push the CH 1 MENU button twice to turn off channel 1, push the CH 2 MENU
button to turn on channel 2, repeat steps 2 and 3. For TDS 224, repeat for CH 3
and CH 4.
2–5
Probe Compensation
Perform this adjustment to match your probe to the input channel. This should be
done whenever you attach a probe for the first time to any input channel.
PROBE
COMP
AUTOSE
T button
1. Set the Probe menu attenuation to 10X. Set the switch to 10X on the P2100
probe and connect the oscilloscope probe to channel 1. Attach the probe tip and
reference lead to the PROBE COMP connectors and then press AUTOSET.
If using the probe hook-tip, ensure a proper connection by firmly twisting the
tip onto the probe.
CH 1
2. Check the shape of the displayed waveform.
Overcompensated
Undercompensated
Compensated correctly
3. If necessary, adjust your probe.
Repeat as necessary.
2–6
Self Calibration
The self calibration routine lets you quickly optimize the oscilloscope signal path
for maximum measurement accuracy. You can run the routine at anytime but you
should always run the routine if the ambient temperature changes by 5_ C or
more.
To compensate the signal path, disconnect any probes or cables from the channel
input connectors. Then, press the UTILITY button and select Do Self Cal to
confirm that you are ready to proceed.
Probe Safety
A guard around the probe body provides a finger barrier for protection from
electric shock.
Finger guard
WARNING. To avoid electric shock when using the probe, keep fingers behind the
guard on the probe body.
To avoid electric shock while using the probe, do not touch metallic portions of
the probe head while it is connected to a voltage source.
Connect the probe to the instrument and connect the ground terminal to ground
before you take any measurements.
Probes are available with various attenuation factors which affect the vertical
scale readout of the oscilloscope.
To change (or check) the probe attenuation setting, press the VERTICAL MENU
button (of the channel you’re using) and then press the menu selection next to
Probe until the correct setting is displayed.
This setting remains in effect until changed again.
2–7
NOTE. The default Probe menu attenuation setting is 10X when the oscilloscope
is shipped.
Be sure that the Attenuation switch on the P2100 probe is set to match the Probe
menu selection in the oscilloscope. The probe switch settings are 1X and 10X.
Attenuation switch
NOTE. When the Attenuation switch is set to 1X, the P2100 probe limits the
bandwidth of the oscilloscope to 7 MHz. To use the full bandwidth of the
oscilloscope, be sure to set the switch to 10X.
Factory Setup
The table below lists the state of the instrument after you select Recall Factory
from the SAVE/RECALL menu.
Control
Sample
Acquire # of averages
16
Channel selection
Channel 1 on
Cursor H Bar position
–3.2 divisions from the center
Cursor V Bar position
–2 divisions from the center
Cursor function
Off
Cursor time units
Seconds
Delayed time base time/div
50 ms/div
Delay time, delayed runs after main
50 ms
Display format
YT
Display graticule type
Full
Display contrast
50%
Display style
Vectors
Display accumulate time
500 ms
Edge trigger coupling
2–8
Setting
Acquire mode
DC
Table 2–1: Factory setup settings (cont.)
Control
Setting
Edge trigger level
0.0 V
Edge trigger slope
Rising
Edge trigger source
Channel 1
Horizontal – main trigger position
50%
Horizontal – time base
Main only
Main time base time/div
500 ms/div
Math waveform function
CH1 + CH2
Saved setups
No change
Saved waveforms
No change
Trigger holdoff
Minimum (500 ns)
Trigger mode
Auto
Trigger type
Edge
Vertical bandwidth (all channels)
Full
Vertical coupling (all channels)
DC
Vertical position (all channels)
0 div
Vertical volts/div. (all channels)
100 mV/div
2–9
2–10
Theory of Operation
This chapter describes the electrical operation of the TDS 200-Series Digital
Real-Time Oscilloscope to the module level. It describes the basic operation of
each functional circuit block shown in figures 3–1 and 3–2.
Main Board
The Main board contains a two or four channel acquisition system, trigger
system, acquisition/display controller, and system processor. Interfaces are
provided for two signal inputs, external trigger input, probe compensation, front
panel, LCD display, extension module, and power supply. Acquisition and
display functions are implemented in two application specific integrated circuits
(ASICs).
Acquisition System
Signals from the CH 1, CH 2, and EXT TRIG, or CH 1, CH 2, CH 3, and CH 4
input connectors pass through attenuators to the acquisition ASIC. The acquisition ASIC contains amplifiers and samplers for each input channel, A/D
converter, and trigger logic. The digitized waveform samples are transferred to
the display ASIC. Low voltage regulators supply isolated power to the acquisition ASIC.
Display System
The display system consists of the display ASIC, DRAM memory, and system
oscillator. Digitized acquisition samples are processed by the display ASIC and
stored into DRAM. The Display ASIC formats the waveform data and generates
drive signals for the LCD display. Front panel scanning is also performed by the
display ASIC.
Processor System
The processor system consists of the microprocessor and memory. The processor
system interprets the front panel controls, programs acquisition and display
parameters, computes waveform measurements, and manages the extension
module interface. Saved setups, waveforms, and calibration constants are stored
in nonvolatile memory. The processor system shares DRAM with the display
system.
Input Signal Interface
BNC connectors are mounted on the main board for the CH 1, CH 2, and trigger
or CH 1, CH 2, CH 3, and CH 4 inputs. The signal inputs are compatible with the
P2100 probes supplied.
Probe Compensation
The PROBE COMP and ground terminals are provided for probe adjustment.
3–1
Theory of Operation
Power Supply
The power supply module generates low voltage DC, a line trigger signal for the
main board, and high voltage AC to power the LCD backlight. The power supply
operates over a wide power line voltage and frequency range.
Display Module
The display module consists of a liquid crystal display (LCD), a display driver,
and a florescent backlight.
Front Panel
The front panel board holds the encoders and push button contacts that the
operator uses to control the instrument. An elastomer mat holds the conductive
pads for the push buttons. Circuits on the front panel convert the matrix into
serial data which is read by the display ASIC on the main Board.
Extension Modules
Each extension module also contains a GPIB, RS–232, and Centronics parallel
printer interface. A ROM on the extension module contains processor instructions to support the communications interfaces. Extension modules interface with
the processor system through the extension module port connector.
3–2
Theory of Operation
Power Supply
3
AC Line
AC Line In
P3
Backlight Power
DC Out
& Line
Trigger
GND
2
J602
Front Panel
A1 Main
7
GND
P53
J131
J603
Display
P131
J90
1
DC In
5
Backlight
J604
J202
Probe Comp
Signal
J902
12
LCD
Display Control
J605
Probe Comp
GND
J51
CH1
J52
Extension
Module
J601
CH2
J1
5
0
J2
9
24
J4
RS232
GPIB
J53
EXT
TRIG
25
J3
CENTRONICS
3–3
Theory of Operation
Power Supply
3
AC Line
AC Line In
P3
Backlight Power
DC Out
& Line
Trigger
GND
2
A1 Main
7
J101
Front Panel
GND
J1
J103
J131
Display
P131
J90
1
DC In
5
J102
J104
Probe Comp
Signal
Backlight
J902
12
LCD
Display Control
J104
Probe Comp
GND
J200
CH1
Extension
Module
J690
J250
J1
5
0
J2
CH2
9
24
J4
J300
25
J3
CH3
RS232
GPIB
CENTRONICS
J350
CH4
3–4
This chapter contains performance verification procedures for the specifications
marked with the n symbol. The following equipment, or a suitable equivalent, is
required to complete these procedures.
Description
Minimum
requirements
DC Voltage Source
17.5 mV to 7 V, ±0.5%
accuracy
Leveled Sine Wave Generator
50 kHz and 100 MHz, ±3%
amplitude accuracy
Examples
Wavetek 9100 Universal
Calibration System with
Oscilloscope Calibration
Module (Option 250)
Time Mark Generator
10 ms period, ±25 ppm
accuracy
Fluke 5500A Multi-product
Calibrator with Oscilloscope
Calibration Option (Option
5500A-SC)
50 W BNC Cable
BNC male to BNC male,
≈ 36 in (1 m) long
Tektronix part number
012-0482-00
50 W BNC Cable
≈ 10 in (25 cm) long
012-0208-00
50 W Feedthrough Termination BNC male and female
connectors
011-0049-01
Dual Banana to BNC Adapter
Banana plugs to BNC female
103-0090-00
BNC T Adapter
BNC male to dual BNC female Tektronix part number
connectors
103-0030-00
4–1
Test Record
Serial
number
Procedure performed by
Date
Test
Passed
Failed
Test result
High limit
Self Test
Oscilloscope tests
Channel 1 DC Gain
Accuracy
Low limit
5 mV/div
33.3 mV
36.7 mV
200 mV/div
1.351 V
1.449 V
2 V/div
13.51 V
14.49 V
5 mV/div
33.3 mV
36.7 mV
200 mV/div
1.351 V
1.449 V
2 V/div
13.51 V
14.49 V
5 mV/div
33.3 mV
36.7 mV
200 mV/div
1.351 V
1.449 V
2 V/div
13.51 V
14.49 V
5 mV/div
33.3 mV
36.7 mV
200 mV/div
1.351 V
1.449 V
2 V/div
13.51 V
14.49 V
Channel 1 Bandwidth
2.12 V
Channel 2 Bandwidth
2.12 V
Channel 3 Bandwidth1
2.12 V
Channel 4 Bandwidth1
2.12 V
Sample Rate and Delay Time Accuracy
–2 divs
+2 divs
Channel 1 Edge Trigger Sensitivity
Stable trigger
Channel 2 Edge Trigger Sensitivity
Stable trigger
Channel 3 Edge Trigger Sensitivity1
Stable trigger
Channel 4 Edge Trigger Sensitivity1
Stable trigger
External Edge Trigger Sensitivity1
Stable trigger
Channel 2 DC Gain
Accuracy
Channel 3 DC Gain
Accuracy1
Channel 4 DC Gain
Accuracy1
1
4–2
Channels 3 and 4 are only on four channel instruments, and External Edge trigger is only on two channel instruments.
Before beginning these procedures, two conditions must first be met:
H
within the operating temperature range specified.
H
You must perform the Self Calibration operation described below. If the
ambient temperature changes by more than 5° C, you must perform the Self
Calibration operation again.
The time required to complete the entire procedure is approximately one hour.
WARNING. Some procedures use hazardous voltages. To prevent electrical shock,
always set voltage source outputs to 0 V before making or changing any
interconnections.
Self Test
Self Calibration
This internal procedure is automatically performed every time the instrument is
powered on. No test equipment or hookups are required. Verify that no error
messages are displayed before continuing with this procedure.
more.
1. Disconnect any probes or cables from the channel 1 and channel 2 input
connectors.
2. Press the UTILITY button and select Do Self Cal to start the routine. The
routine takes approximately one minute to complete.
3. Verify that Self calibration passed.
4–3
Check DC Gain Accuracy
This test checks the DC gain accuracy of all input channels.
1. Set the DC voltage source output level to 0 V.
2. Set up the instrument using the following steps:
Press menu button
Select menu item
Select setting
SAVE/
RECALL
Recall
Factory
CH 1
Probe
1X
ACQUIRE
Average
16
MEASURE
Source
An unchecked channel
Type
Mean
3. As shown below, connect the instrument channel selected in the table to the
DC voltage source.
Digitizing oscilloscope
DC voltage
source
Dual banana
to BNC
adapter
BNC cable
4. For each VOLTS/DIV setting listed below, perform the following steps:
a. Set the DC voltage source output level to the positive voltage listed and
then record the mean measurement as Vpos.
b. Reverse the polarity of the DC voltage source and then record the mean
measurement as Vneg.
c. Calculate Vdiff = Vpos – Vneg and then compare Vdiff to the accuracy
limits in the table.
VOLTS/DIV setting
Accuracy limits for Vdiff
5 mV/div
+17.5 mV, –17.5 mV
33.3 mV to 36.7 mV
200 mV/div
+700 mV, –700 mV
1.351 V to 1.449 V
2 V/div
4–4
DC voltage source output levels
+7.00 V, –7.00 V
13.51 V to 14.49 V
5. Set DC voltage source output level to 0 V.
6. Disconnect the test setup.
7. Repeat steps 1 through 6 until all input channels have been checked.
Check Bandwidth
This test checks the bandwidth of all input channels.
1. Set up the instrument using the following steps:
Press menu button
Select menu item
Select setting
SAVE/
RECALL
Recall
Factory
CH 1
Probe
1X
ACQUIRE
Average
16
TRIGGER
Coupling
Noise Reject
MEASURE
Source
Type
Pk-Pk
2. As shown below, connect the instrument channel selected in the table to the
leveled sine wave generator.
Leveled
sine wave
generator
Output
BNC cable
50 W feedthrough
terminator
3. Set the instrument VOLTS/DIV to 500 mV/div.
4. Set the instrument SEC/DIV to 10 ms/div.
5. Set the leveled sine wave generator frequency to 50 kHz.
6. Set the leveled sine wave generator output level so the peak-to-peak
measurement is between 2.98 V and 3.02 V.
7. Set the leveled sine wave generator frequency to 60 MHz if you are checking
a TDS 210 or to 100 MHz if you are checking a TDS 220 or TDS 224.
4–5
8. Set the instrument SEC/DIV to 10 ns/div.
9. Check that the peak-to-peak measurement is ≥2.12 V.
10. Disconnect the test setup.
11. Repeat steps 1 through 10 until all input channels have been checked.
Check Sample Rate and
Delay Time Accuracy
This test checks the time base accuracy.
Press menu button
Select menu item
Select setting
SAVE/
RECALL
Recall
Factory
CH 1
Probe
1X
2. Connect the instrument to the time mark generator as shown below.
Time mark
generator
Output
50 W feedthrough
terminator
BNC cable
3. Set the time mark generator period to 10 ms.
4. Set the instrument VOLTS/DIV to 500 mV/div.
5. Set the instrument Main SEC/DIV to 1 ms/div.
6. Press SET LEVEL TO 50%.
7. Use the vertical POSITION control to center the test signal on screen.
8. Change the instrument setup using the following steps:
Press menu button
Select setting
HORIZONTAL
4–6
Select menu item
Window
9. Set the instrument Window SEC/DIV to 500 ms/div.
NOTE. Setting the Window SEC/DIV to a setting slower than the Main SEC/DIV
setting resets the main timebase. If this happens, reset the Main SEC/DIV and
then the Window SEC/DIV settings.
10. Use the horizontal POSITION control to set the window position to
10.00 ms.
11. Set the instrument Window SEC/DIV to 500 ns/div.
12. Check that the rising edge of the marker crosses the center horizontal
graticule line within ±2 divisions of center graticule.
NOTE. One division of displacement from graticule center corresponds to
a 50 ppm time base error.
13. Disconnect the test setup.
Check Edge Trigger
Sensitivity
This test checks the edge trigger sensitivity for all input channels.
Press menu button
Select menu item
Select setting
SAVE/
RECALL
Recall
Factory
CH 1
Probe
1X
TRIGGER
Mode
Normal
ACQUIRE
Sample
MEASURE
Source
Type
Pk-Pk
4–7
Leveled
sine wave
generator
Output
50 W feedthrough
terminator
BNC cable
3. Set the leveled sine wave generator frequency to 60 MHz if you are checking
5. Set the instrument SEC/DIV to 10 ns/div.
6. Set the leveled sine wave generator output level to approximately 750 mVp-p
so that the measured amplitude is approximately 750 mV. (The measured
amplitude can fluctuate around 750 mV.)
7. Press SET LEVEL TO 50%. Adjust TRIGGER LEVEL as necessary and
then check that triggering is stable.
Press menu button
Select menu item
Select setting
TRIGGER
Slope
Falling
9. Press SET LEVEL TO 50%. Adjust TRIGGER LEVEL as necessary and
10. Change the instrument setup using the following steps:
Press menu button
Select menu item
Select setting
TRIGGER
Slope
Rising
11. Disconnect the test setup.
12. Repeat steps 1 through 11 until all input channels have been checked.
4–8
Check External Edge
Trigger Sensitivity
TDS 210 and TDS 220 only: this test checks the edge trigger sensitivity for the
external trigger.
Press menu button
Select menu item
Select setting
SAVE/
RECALL
Recall
Factory
CH 1
Probe
1X
TRIGGER
Source
Ext
Mode
Normal
ACQUIRE
Sample
MEASURE
Source
CH1
Type
Pk-Pk
2. Connect the instrument to the leveled sine wave generator as shown below.
Leveled
sine wave
generator
50 W
feedthrough
terminator
Output
BNC
cable
BNC cable
BNC T connector
a TDS 210 or to 100 MHz if you are checking a TDS 220.
4–9
Press menu button
Select menu item
Select setting
TRIGGER
Slope
Falling
Press menu button
Select menu item
Select setting
TRIGGER
Slope
Rising
4–10
This chapter contains adjustment procedures for the TDS 200-Series Digital
Real-Time Oscilloscope.
Only qualified personnel should perform service procedures. Read the Service
procedures. Also refer to the chapter Operating Information for information
about using the TDS 200s.
NOTE. The voltage references inside the TDS 200s are very stable over time and
should not require routine updates. Before performing any procedure in this
chapter, first verify that the instrument does not meet specifications. Refer to the
chapter Performance Verification to verify the specifications.
Required Equipment
The following equipment, or a suitable equivalent, is required to complete these
procedures.
Description
Minimum requirements
DC Voltage Source
–20 V to 20 V, ±0.1% accuracy Wavetek 9100 Universal Calibration
System with Oscilloscope
Calibration Module (Option 250)
Leveled Sinewave
Generator
50 kHz and 100 MHz, ±0.6%
amplitude accuracy
5500A-SC)
50 W BNC Cable
≈ 36 in (1 m) long
Tektronix part number 012-0482-00
50 W BNC Cable (two)
Tektronix part number 012-0208-00
50 W Feedthrough
Termination
BNC male and female
connectors
Tektronix part number 011-0049-01
Dual Banana to BNC
Adapter
Tektronix part number 103-0090-00
BNC T
One male and two female
BNC connectors
Tektronix part number 103-0030-00
Examples
5–1
The adjustment procedure consists of 55 steps. Each step requires an external
voltage source for new calibration constants.
Before performing the adjustment procedure, you must warm up the instrument
for at least ten minutes in an ambient temperature between 20° C and 30° C.
Adjustments performed prior to warm-up or outside this temperature range may
result in poor performance.
If all steps in the procedure are completed successfully, a “Pass” message is
displayed and the new calibration constants take affect. If any step fails, the
procedure is aborted and the current calibration is not affected. You can choose to
abort the procedure at any step by selecting FCAL ABORT without affecting the
current calibration.
The equipment setups, shown in Figure 5–2, are required to complete the
adjustment procedure. Tables 5–2 and 5–3 list the steps in the procedure and the
signal requirement for each step.
Enable the Service Menu
You must enable the service menu to perform the adjustment procedure. With the
instrument on, insert a narrow object such as a plastic alignment tool into the
access hole, shown in Figure 5–1, until you feel the click of the momentary
button. Hold this button in and press the front panel UTILITY button.
Access to
service menu
enable button
A Service menu selection replaces the Language menu selection, giving you
access to the service routines.
After adjustment is complete, disable the service menu by again holding in the
internal momentary button and pressing the front panel UTILITY button.
5–2
DC voltage setup 1
Four Channel
Two Channel
DC voltage
source
Dual banana to
BNC adapter
10 inch
BNC cables
BNC cable
10 inch
BNC cables
BNC ‘T’ connector
10 inch
BNC cable
BNC cable
10 inch
BNC cables
BNC ‘T’
connector
s
BNC ‘T’
connector
10 inch BNC cable
DC voltage setup 2
DC voltage
source
Dual banana
to BNC
adapter
BNC cable
Sinewave generator setup
Leveled
sine wave
generator
Edge setup
Four Channel
Square wave
generator
Output
Output
50 W feedthrough
terminator
(Table 5–2 and 5–3 list
proper channel)
BNC cable
50 W feedthrough
terminator
proper channel)
BNC cable
5–3
Use this procedure to load new calibration constants.
1. Turn on the instrument and let it warm up for at least ten minutes.
2. Enable the Service menu as described on Page 5–2.
3. Press the UTILITY button and select Service from the menus.
4. Select FCAL INIT to initiate the factory calibration routine. You are now in
step 1 of the procedure.
5. As indicated by Step 1 in Table 5–2 on page 5–5 (TDS 210 and TDS 220)
or Table 5–3 on page 5–7 (TDS 224), connect a 0.0 VDC source to the
channel inputs.
6. Press FCAL STEP to load the calibration constant for step 1.
7. Follow the screen prompts along with Table 5–2 on page 5–5 (TDS 210 and
TDS 220) or Table 5–3 on page 5–7 (TDS 224) for the remainder of the
routine, applying the proper signal for each step.
NOTE. During some steps, the instrument may appear to be idle for several
minutes while it is processing information internally. Please have patience.
If any step in the procedure fails, the procedure terminates. Failure will result if
an incorrect signal source is connected. If this happens, you must start the
procedure over from the beginning.
If the procedure completes successfully, a “Pass” message is displayed and the
new adjustment takes effect.
8. Compensate the signal path by performing the self calibration routine.
Disconnect any probes or cables from the channel input connectors. Then,
press the UTILITY button and select Do Self Cal to confirm that you are
ready to proceed.
9. Disable the Service menu option.
5–4
Step:
Firmware
Below V 2.00
Step
Firmware
V 2.00 and Up
Equipment setup
Signal source setting
Input
1
1
DC Voltage 1
0.0 V
CH1 & CH2
2
2
DC Voltage 1
–20.0 V
3
3
DC Voltage 1
–2.0 V
4
4
DC Voltage 1
–1.6 V
5
5
DC Voltage 1
–1.2 V
6
6
DC Voltage 1
–0.8 V
7
7
DC Voltage 1
–0.4 V
8
8
DC Voltage 1
–0.32 V
9
9
DC Voltage 1
–0.2 V
10
10
DC Voltage 1
–0.16 V
11
11
DC Voltage 1
–0.08 V
12
12
DC Voltage 1
–0.04 V
13
13
DC Voltage 1
–0.03 V
14
14
DC Voltage 1
–0.02 V
15
15
DC Voltage 1
–0.015 V
16
16
DC Voltage 1
0.015 V
17
17
DC Voltage 1
0.02 V
18
18
DC Voltage 1
0.03 V
19
19
DC Voltage 1
0.04 V
20
20
DC Voltage 1
0.08 V
21
21
DC Voltage 1
0.16 V
22
22
DC Voltage 1
0.2 V
23
23
DC Voltage 1
0.32 V
24
24
DC Voltage 1
0.4 V
25
25
DC Voltage 1
0.8 V
26
26
DC Voltage 1
1.2 V
27
27
DC Voltage 1
1.6 V
28
28
DC Voltage 1
2.0 V
29
29
DC Voltage 1
20.0 V
30
30
DC Voltage 1
0.0 V
31
31
DC Voltage 2
–5.0 V
32
32
DC Voltage 2
–1.0 V
33
33
DC Voltage 2
1.0 V
34
34
DC Voltage 2
5.0 V
EXT TRIG
5–5
Table 5–2: TDS 210 and TDS 220 Adjustment steps (cont.)
Step:
Firmware
Below V 2.00
Step
Firmware
V 2.00 and Up
Equipment setup
Input
35
35
Sinewave Generator, 50 Ohm
5 kHz, 1 Vp-p
CH1
36
36
1 MHz, 1 Vp-p
37
37
50 kHz, 1 Vp-p
38
38
20 MHz, 1 Vp-p
39
39
TDS 210: 60 MHz, 1 Vp-p
TDS 220: 100 MHz, 1 Vp-p
40
40
5 kHz, 2.5 Vp-p
41
41
1 MHz, 2.5 Vp-p
42
42
50 kHz, 2.5 Vp-p
43
43
20 MHz, 2.5 Vp-p
44
44
TDS 210: 60 MHz, 2.5 Vp-p
TDS 220: 100 MHz, 2.5 Vp-p
45
45
None
46
Edge, 50 Ohm
1 kHz, 0 to –800 mV
46
47
5 kHz, 1 Vp-p
47
48
1 MHz, 1 Vp-p
48
49
50 kHz, 1 Vp-p
49
50
20 MHz, 1 Vp-p
50
51
CH2
51
52
5 kHz, 2.5 Vp-p
52
53
1 MHz, 2.5 Vp-p
53
54
50 kHz, 2.5 Vp-p
54
55
20 MHz, 2.5 Vp-p
55
56
5–6
Step
Equipment setup
Input
1
DC Voltage 1
0.0 V
2
DC Voltage 1
–20.0 V
CH1, CH2, CH3
& CH4
3
DC Voltage 1
–2.0 V
4
DC Voltage 1
–1.6 V
5
DC Voltage 1
–1.2 V
6
DC Voltage 1
–0.8 V
7
DC Voltage 1
–0.4 V
8
DC Voltage 1
–0.32 V
9
DC Voltage 1
–0.2 V
10
DC Voltage 1
–0.16 V
11
DC Voltage 1
–0.08 V
12
DC Voltage 1
–0.04 V
13
DC Voltage 1
–0.03 V
14
DC Voltage 1
–0.02 V
15
DC Voltage 1
–0.015 V
16
DC Voltage 1
0.015 V
17
DC Voltage 1
0.02 V
18
DC Voltage 1
0.03 V
19
DC Voltage 1
0.04 V
20
DC Voltage 1
0.08 V
21
DC Voltage 1
0.16 V
22
DC Voltage 1
0.2 V
23
DC Voltage 1
0.32 V
24
DC Voltage 1
0.4 V
25
DC Voltage 1
0.8 V
26
DC Voltage 1
1.2 V
27
DC Voltage 1
1.6 V
28
DC Voltage 1
2.0 V
29
DC Voltage 1
20.0 V
30
DC Voltage 1
0.0 V
5–7
Table 5–3: TDS 224 Adjustment steps (cont.)
Step
Input
31
5 kHz, 1 Vp-p
CH1
32
1 MHz, 1 Vp-p
33
50 kHz, 1 Vp-p
34
20 MHz, 1 Vp-p
35
100 MHz, 1 Vp-p
36
5 kHz, 2.5 Vp-p
37
1 MHz, 2.5 Vp-p
38
50 kHz, 2.5 Vp-p
39
20 MHz, 2.5 Vp-p
40
100 MHz, 2.5 Vp-p
41
77.133 MHz, 1 Vp-p
42
Edge, 50 Ohm
1 kHz, 0 to –800 mV
43
5 kHz, 1 Vp-p
44
1 MHz, 1 Vp-p
45
50 kHz, 1 Vp-p
46
20 MHz, 1 Vp-p
47
100 MHz, 1 Vp-p
48
5 kHz, 2.5 Vp-p
49
1 MHz, 2.5 Vp-p
50
50 kHz, 2.5 Vp-p
51
20 MHz, 2.5 Vp-p
52
100 MHz, 2.5 Vp-p
53
5 kHz, 1 Vp-p
54
1 MHz, 1 Vp-p
55
50 kHz, 1 Vp-p
56
20 MHz, 1 Vp-p
57
100 MHz, 1 Vp-p
58
5 kHz, 2.5 Vp-p
59
1 MHz, 2.5 Vp-p
60
50 kHz, 2.5 Vp-p
61
20 MHz, 2.5 Vp-p
62
100 MHz, 2.5 Vp-p
63
77.133 MHz, 1 Vp-p
64
5–8
Equipment setup
Edge, 50 Ohm
1 kHz, 0 to –800 mV
CH2
CH3
Step
Equipment setup
Input
65
5 kHz, 1 Vp-p
CH4
66
1 MHz, 1 Vp-p
67
50 kHz, 1 Vp-p
68
20 MHz, 1 Vp-p
69
100 MHz, 1 Vp-p
70
5 kHz, 2.5 Vp-p
71
1 MHz, 2.5 Vp-p
72
50 kHz, 2.5 Vp-p
73
20 MHz, 2.5 Vp-p
74
100 MHz, 2.5 Vp-p
5–9
5–10
Maintenance
This chapter contains the information needed to do periodic and corrective
maintenance on the oscilloscope. The following sections are included:
H
Preparation refers you to important information contained elsewhere in this
manual.
H
Preventing ESD provides general information on preventing damage to
internal modules when doing maintenance.
H
Inspection and Cleaning provides procedures for inspecting the instrument
and cleaning its external and internal modules.
H
Removal and Installation Procedures provide procedures for the removal and
installation of modules.
H
Troubleshooting provides information for isolating failed modules.
H
Repackaging Instructions provides packaging information for shipment or
storage.
Preparation
about using the oscilloscope.
Preventing ESD
When performing any service which requires internal access to the instrument,
adhere to the following precautions to avoid damaging internal modules and their
components due to electrostatic discharge (ESD).
CAUTION. Static discharge can damage any semiconductor component in this
instrument.
1. Minimize handling of static-sensitive modules.
2. Transport and store static-sensitive modules in their static protected
containers or on a metal rail. Label any package that contains static-sensitive
modules.
6–1
Maintenance
3. Discharge the static voltage from your body by wearing a grounded antistatic
wrist strap while handling these modules. Do service of static-sensitive
modules only at a static-free work station.
4. Nothing capable of generating or holding a static charge should be allowed
on the work station surface.
5. Handle circuit boards by the edges when possible.
6. Do not slide the modules over any surface.
7. Avoid handling modules in areas that have a floor or work-surface covering
capable of generating a static charge.
This section describes how to inspect for dirt and damage, and how to clean the
instrument. Inspection and cleaning are done as preventive maintenance.
Preventive maintenance, when done regularly, may prevent malfunctions and
enhance reliability.
Preventive maintenance consists of visually inspecting and cleaning the
instrument, and using general care when operating it. How often to do maintenance depends on the severity of the environment in which the instrument is
used. A proper time to perform preventive maintenance is just before adjustment
of the instrument.
General Care
Protect the instrument from adverse weather conditions. The instrument is not
waterproof.
Do not store or leave the instrument where the liquid crystal display (LCD) will
be exposed to direct sunlight for long periods of time.
The front and back cases help keep dust out of the instrument and must be in
place during normal operation.
CAUTION. To avoid damage to the instrument or probes, do not expose it to any
sprays, liquids, or solvents.
6–2
Maintenance
Procedures
Inspect and clean the instrument and probes as often as operating conditions
require. The collection of dirt on internal components can cause them to overheat
and breakdown. Dirt also provides an electrical conduction path that could cause
an instrument failure, especially under high-humidity conditions.
Inspection — Exterior. Inspect the outside of the instrument for damage, wear, and
missing parts. Instruments that appear to have been dropped or otherwise abused
should be checked thoroughly to verify correct operation and performance.
Immediately repair defects that could cause personal injury or lead to further
damage to the instrument.
Cleaning Procedure — Exterior. To clean the instrument exterior and probes,
1. Remove loose dust on the outside of the instrument and probes with a lint
free cloth. Use care to avoid scratching the clear plastic display shield.
2. Use a soft cloth dampened with water to clean the instrument. You can use a
75% isopropyl alcohol solution for more efficient cleaning.
CAUTION. To avoid damage to the surface of the instrument or probes, do not use
any abrasive or chemical cleaning agents.
Inspection — Interior. During normal use, the instrument case guards the internal
components against common dust and dirt. However, if it appears that the
instrument may have been damaged or has been exposed to an unusually dirty
environment, you should inspect and clean the interior.
To inspect and clean the interior, disassemble the instrument by completing each
of the disassembly procedures in the order given starting on page 6–12.
Inspect the internal portions of the instrument for damage and wear using
Table 6–1 as a guide. Defects found should be repaired immediately.
Item
Inspect for
Repair action
Front and Back
Case
Cracks or deformations.
Scratched lettering or display
filter. Loose connectors or
labels.
Repair or replace defective hardware.
Gasket and Foam
Pads
Misplaced or missing pieces.
Relocate or replace defective pieces.
6–3
Maintenance
Table 6–1: Internal inspection check list (cont.)
Item
Inspect for
Repair action
Circuit boards
Loose, broken, or corroded
Remove failed module and replace with a
solder connections. Burned
new module.
circuit boards. Burned, broken,
or cracked circuit-run plating.
Solder connections Cold solder or rosin joints.
Resolder joint and clean with isopropyl
alcohol.
Wiring and cables
Firmly seat connectors. Repair or replace
modules with defective wires or cables.
Loose plugs or connectors.
Burned, broken, or frayed
wiring.
Cleaning Procedure — Interior. To clean the instrument interior, perform the
following steps:
1. Blow off dust with dry, low-pressure, deionized air (approximately 9 psi).
2. Remove any remaining dust with a lint-free cloth dampened in isopropyl
alcohol (75% solution) and rinse with warm deionized water. (A cottontipped applicator is useful for cleaning in narrow spaces and on circuit
boards.)
STOP. If, after performing steps 1 and 2, a module is clean upon inspection, skip
steps 3 and 4. Proceed with step 5.
If steps 1 and 2 do not remove all the dust or dirt, all modules in the instrument
except the Display Module may be spray washed using a solution of 75%
isopropyl alcohol. Proceed with step 3.
3. Spray wash dirty parts with the 75% isopropyl alcohol solution and wait
60 seconds for the majority of the alcohol to evaporate.
4. Use hot (49_ C to 60_ C / 120_ F to 140_ F) deionized water to thoroughly
rinse the parts.
5. Dry all parts with low-pressure, deionized air.
6. Dry all components and assemblies in an oven or drying compartment using
low-temperature (52_ C to 66_ C / 125_ F to 150_ F) circulating air.
Lubrication. There is no periodic lubrication required for the oscilloscope.
6–4
Maintenance
This section describes how to remove and install the major mechanical and
electrical modules. It provides the following information:
H
H
List of Modules on page 6–6 describes where to locate a list of replaceable
modules.
H
Summary of Procedures on page 6–7 lists the procedures for removal and
H
Preparation
Tools Required on page 6–6 describes the tools needed to perform the
procedures.
Rear Feet, on page 6–8, begins a series of procedures that describe the
removal and installation of modules.
Please read the following warning statement. Then read the following general
instructions before removing a module.
WARNING. Only qualified personnel should perform service procedures. Before
doing this or any other procedure in this manual, read the General Safety
Summary and Service Safety Summary found at the beginning of this manual.
Always disconnect the power cord before disassembling any portion of the
instrument.
Also, to prevent damage to electrical components, read Preventing ESD on
page 6–1.
1. Use Figure 10–1 (the exploded view of the instrument) on page 10–8 to give
you an overview of how the entire instrument is assembled.
2. Read the Summary of Procedures on page 6–7 to understand which
procedure to use to access a given module. Then read Tools Required on
page 6–6 for a list of tools needed to remove and install modules in the
oscilloscope.
3. If you are disassembling the oscilloscope for cleaning, refer to the Inspection
and Cleaning procedure on page 6–3 for cleaning instructions.
6–5
Maintenance
Tools Required
You will need the following tools to remove and replace all the modules in the
oscilloscope.
Tools
Torque-limiting screwdriver, 12 in lb (1.3 N ⋅ m) range with Torx T-15 tip
Screwdriver with !@4 inch flat blade
Pliers
Sharp knife (X-ACTO knife or equivalent)
List of Modules
6–6
The Mechanical Parts List chapter provides a list of all replaceable modules.
Any electrical or mechanical module, assembly, or part listed in the parts list is
referred to as a module.
Maintenance
Table 6–2 lists all procedures in order of increasing complexity. Complete
disassembly is best achieved by following the procedures in the order given.
Procedure
Begins on Page
Rear Feet
6–8
Front-Case Label
6–10
Rear-Case Label
6–11
Handle
6–12
Power Button
6–13
Rear Case
6–14
Front Feet
6–16
EMI Clips
6–17
Flip Stand
6–18
Power Line Fuse
6–20
Power Supply Module
6–21
Front Panel Knobs
6–23
Internal Assembly
6–24
Main Board Module
6–26
Display Module
6–28
Front Panel Module
6–30
Keypad
6–33
Display Shield
6–35
Front Case
6–37
TDS 200 Series Digital Oscilloscope Service Manual
6–7
Maintenance
Rear Feet
No tools are required to remove the feet.
Removal. Use this procedure to remove the rear feet.
1. Place the instrument face down on a soft surface (such as an anti-static mat).
2. Slide the feet up and out of the rear case as shown in Figure 6–1.
NOTE. You may want to use a flat blade screwdriver to pry up on the foot. Use
care not to mar the surface of the rear case.
Rear foot (2)
6–8
Maintenance
Installation. Use this procedure to install the feet.
1. Place the rear case face down on a soft surface (such as an anti-static mat).
2. Align the slots of the rear foot with the guides in the rear case as shown in
Figure 6–2 and press the foot in.
Rear foot
(2)
6–9
Maintenance
Front-Case Label
You will need a sharp knife and needle-nose pliers to remove the front-case label.
CAUTION. Removal of a front-case label will destroy it. Do not remove it unless it
needs to be replaced.
Removal. Use the sharp knife to pry up a corner of the label and use the pliers to
peel the label off.
Installation. Use this procedure to install a new front-panel label in either a new
or the existing front case.
1. Peel the paper off the back side of the new label.
2. Handling the label by its edges only, align it with the recess in the front case
as shown in Figure 6–3.
3. When properly aligned, press the new label into place.
6–10
Maintenance
Rear-Case Label
You will need a sharp knife and needle-nose pliers to remove the rear-case label.
CAUTION. Removal of a rear-case label will destroy it. Do not remove it unless it
peel the label off.
Installation. Use this procedure to install a new rear-case label in either a new or
the existing rear case.
2. Handling the label by its edges only, align it with the recess in the rear case
as shown in Figure 6–4.
Rear label
6–11
Maintenance
Handle
You will need a flat-blade screw driver to remove the handle.
Removal. Use this procedure to remove the handle.
1. Place the instrument on its feet.
2. With the handle laying down, pry one end of the handle out of the case as
shown in Figure 6–5.
NOTE. Use care not to mar the surface with the screwdriver.
3. Slide the handle towards the back of the rear case.
6–12
Maintenance
Installation. Use this procedure to install the handle.
2. From behind the rear case with the handle laying down, slide the handle
along the groves in the case until the handle snaps into place as shown in
Figure 6–6.
Figure 6–6: Installing the handle
Power Button
Remove the power button by firmly grasping the button (with pliers if necessary)
and pull it straight out of the cabinet. Reinstall the button by pressing it onto the
power button shaft until it snaps into place.
NOTE. Wrap the plier jaws with tape (such as electrical) to avoid marring the
button surface.
6–13
Maintenance
Rear Case
You will need a torque-limiting Torx T-15 screwdriver and a flat-blade
screwdriver for this procedure.
Removal. First remove the handle and power button using the procedures on
pages 6–12 and 6–13, then use this procedure to remove the rear case.
1. If installed, remove the extension module from the rear of the instrument.
CAUTION. Do not remove the rear case with an extension module installed.
2. Place the instrument face down on a soft surface (such as an anti-static mat).
3. Remove the two screws located in the handle recesses. Refer to Figure 6–7.
4. Slip the flat-blade screwdriver between the rear case and front case at the
bottom of the instrument to disengage the tabs and lift the rear case off.
Installation. Use this procedure to install the rear case.
2. Slide the rear case over the assembly until the bottom tabs of the front case
lock into the rear case.
CAUTION. Be careful to align the guide ribs and to not pinch any wires or other
parts between the front and rear cases. If used, make sure that the copper mesh
grounding tube is in the proper place as shown in Figure 6–7.
3. Install the two screws located in the handle recesses.
CAUTION. Avoid cross threading the screws that secure the rear case to the front
case. An easy way to avoid this is to slip the screw into the hole and, using slight
pressure, slowly rotate the screw counterclockwise until you feel the screw slip
into the screw threads. Then tighten the screw clockwise. A properly started
screw will have very little resistance going in.
4. Install the handle and the power button.
6–14
Maintenance
Retaining screw
Rear case
Retaining screw
Retaining tab (7)
Copper mesh tubing
Front case
Figure 6–7: Removing and installing the rear cover
6–15
Maintenance
Front Feet
Removal. First remove the handle, power button, and rear case using the
procedures on pages 6–12, 6–13, and 6–14. Then, slide the front feet out of the
front case as shown in Figure 6–8.
Installation. Slide the feet into the slots provided in the front case as shown in
Figure 6–8. Use the installation procedures for each module removed to
reassemble the instrument.
Front foot (2)
Front case
Figure 6–8: Removing and installing the front feet
6–16
Maintenance
EMI Clips
procedures on pages 6–12, 6–13, and 6–14. Then, slide the four EMI clips out of
the rear case as shown in Figure 6–9.
NOTE. The EMI clips can be pushed out from the rear of the case using a narrow
flat-blade screwdriver.
Installation. Slide the four EMI clips into the slots provided in the rear case as
shown in Figure 6–9. Use the installation procedures for each module removed to
Figure 6–9: Removing and installing the EMI clips
6–17
Maintenance
Flip Stand
procedures on pages 6–12, 6–13, and 6–14. Then, use the following procedure to
remove the flip stand.
2. With the stand folded against the rear case (in the up position), push the stand
out of the rear case channels as shown in Figure 6–10.
Rear case
Flip stand
Figure 6–10: Removing the flip stand
6–18
Maintenance
Installation. Use this procedure to install the flip stand.
1. Orient the flip stand to the rear case, aligning the flip stand slots to the rear
case tabs as shown in Figure 6–11.
2. With the stand in the folded up position, snap the stand into place.
3. Use the installation procedures for each module removed to reassemble the
instrument.
Rear case
Flip stand
Figure 6–11: Installing the flip stand
6–19
Maintenance
Power Line Fuse
First remove the handle, power button, and rear case using the procedures on
pages 6–12, 6–13, and 6–14 to access the line fuse.
The line fuse is located on the power supply module as shown in Figure 6–12. If
the fuse tests bad, replace it with a fuse of the same type and rating.
instrument.
Line fuse
Power supply module
Front case
Figure 6–12: Line fuse location
6–20
Maintenance
Power Supply Module
remove the power supply module. Refer to Figure 6–13.
1. Disconnect the following wires.
a. The line cord ground wire on the main board at J602 or J101.
b. The seven-conductor ribbon cable on the main board at J131.
c. The two-conductor backlight cable on the power supply module.
2. Slightly bend back the two securing tabs at the bottom of the power supply
module and lift the module out of the inner chassis.
Two-conductor
backlight cable
Seven-conductor
ribbon cable
Power supply module
Front case
Line cord
ground wire
Figure 6–13: Removing the power supply module
6–21
Maintenance
Installation. Use this procedure to install the power supply module. Refer to
Figure 6–14.
1. Place the power supply module into the inner chassis, placing the board
under the three tabs at the top of the inner chassis.
2. Snap the bottom of the module into place.
3. Reconnect the following wires.
4. Use the installation procedures for each module removed to reassemble the
instrument.
Two-conductor
backlight cable
Seven-conductor
ribbon cable
Power supply module
Front case
Line cord
ground wire
Figure 6–14: Installing the power supply module
6–22
Maintenance
Front-Panel Knobs
Remove the front panel knobs by firmly grasping the knob (with pliers if
necessary) and pull it away from the front panel. Install front panel knobs by
aligning the keyed knob with the shaft and pressing the knob onto the shaft.
knob surface.
6–23
Maintenance
Internal Assembly
remove the internal assembly.
1. Remove all the front panel knobs.
2. Lift the entire internal assembly (housing all of the boards and the display
screen) out of the front case as shown in Figure 6–15.
NOTE. The switch keypad will most likely remain inside the front case. It does not
need to be removed with the internal assembly.
Internal assembly
Keypad
Conductive foam pad
Front case
Figure 6–15: Removing and installing the internal assembly
6–24
Maintenance
Installation. Use this procedure to install the internal assembly.
1. Place the internal assembly into the front case as shown in Figure 6–15.
2. If used, place the copper mesh grounding tube between the main board
module shield and the front case as shown in Figure 6–16. This part is not
used on later instruments; check the Mechanical Parts List for your serial
number (see page 10–8).
instrument.
Front case
Copper
mesh tubing
Figure 6–16: Installing the copper mesh grounding tube
6–25
Maintenance
Main Board Module
remove the main board module.
1. Remove the entire internal assembly from the front case using the procedure
on page 6–24.
2. Disconnect the following wires on the main board. Refer to Figure 6–17.
a. The line cord ground wire at J602 or J101 and the seven-conductor
ribbon cable at J131 from the power supply module.
b. The ribbon cable at J603 or J103 from the front panel module.
c. The ribbon cable at J202 or J102 from the display module.
3. From the bottom of the internal chassis, slightly bend the two securing tabs
and tilt the board down until it clears the tabs. Refer to Figure 6–17.
4. Pull the board out of the front slots of the internal chassis (near the BNC
connectors).
6–26
Maintenance
Line cord ground wire
Power supply
ribbon cable
Internal chassis
Display
module ribbon
cable
Front panel cable
Main
board
Figure 6–17: Main board removal
Installation. Use this procedure to install the main board module. Refer to
Figure 6–18.
1. Working from the bottom of the internal chassis, place the front of the main
board module into the slots on the internal chassis (near the BNC connectors).
2. Keeping the module in place, press the board into place so that the securing
tabs lock onto the bottom of the board. Note that there are two board guides
on the chassis that must match the cutouts on the main board.
3. Reconnect the following wires on the main board. Refer to Figure 6–17.
4. Place the entire internal assembly (housing all of the boards and the display
screen) into the front case as shown in Figure 6–16.
6–27
Maintenance
5. Use the installation procedures for each module removed to reassemble the
instrument.
Power supply
ribbon cable
Internal chassis
Display
module ribbon
cable
Front panel cable
Board guides
Cutout
Main
board
Figure 6–18: Main board installation
Display Module
remove the display module.
on page 6–24.
2. Disconnect the display module ribbon cable at J202 or J102 on the main
board by pulling straight up from the connector. Refer to Figure 6–17.
3. Disconnect the backlight cable from the power supply module. Refer to
Figure 6–14.
4. Slightly bend the two securing tabs on the left side of the display module and
lift the left side up until it clears the tabs. Then slide the display to the left
until the right-side tabs are disengaged. Refer to Figure 6–19.
6–28
Maintenance
CAUTION. The display shield is attached to the display module. Removing the
shield will destroy the attaching gasket.
Internal chassis
Backlight
cable
Display module
Ribbon cable
Figure 6–19: Removing the display module
Installation. Use this procedure to install the display module.
1. Route the display ribbon cable through the opening in the internal chassis as
shown in Figure 6–20. If installing a new module, fold the ribbon cable as
shown.
2. Place the tabs on the right side of the display module into the slots of the
internal chassis located under the front panel board. It is not necessary to
remove the front panel board. Refer to Figure 6–20.
3. Route the backlight cable through the opening in the internal chassis as
4. Lower the left side of the display module into the internal chassis until the
two securing tabs lock on to the display module. Note that the backlight
cable protruding from each end of the module must align with the cutouts of
the inner chassis. Refer to Figure 6–20.
5. Reconnect the display module ribbon cable at J202 or J102 on the main
board by pushing the cable straight down into the connector. Refer to Figure
6–17.
6–29
Maintenance
6. Reconnect the backlight cable to the power supply module. Refer to
Figure 6–14.
7. Place the entire internal assembly into the front cover. Refer to Figure 6–16.
8. Use the installation procedures for each module removed to reassemble the
instrument.
Backlight cable to
power supply
module
Internal chassis
Display module
Ribbon cable to
main board
Do not pinch backlight cable
when installing the display module.
Left side view of
internal chassis
Figure 6–20: Installing the display module
Front Panel Module
remove the front panel module.
on page 6–24.
2. Disconnect the front panel ribbon cable at J200 on the main board. Refer to
Figure 6–17.
6–30
Maintenance
3. Disengage the two securing tabs at the bottom of the board and slide the
board down and out of the internal chassis. Refer to Figure 6–21
Internal chassis
Front panel module
Figure 6–21: Removing the front panel module
6–31
Maintenance
Installation. Use this procedure to install the front panel module.
1. Route the front panel ribbon cable through the opening in the chassis as
shown in Figure 6–22.
2. Insert the top of the front panel module under the tabs at the top of the
chassis and press the board down until the securing tabs at the bottom of the
chassis lock into place. Refer to Figure 6–21.
3. Reconnect the front panel ribbon cable at J603 or J103 on the main board.
4. Place the entire internal assembly into the front case as shown in Figure 6–15.
instrument.
Internal chassis
Ribbon cable to
J202 on main
board
Front panel module
Figure 6–22: Installing the front panel module
6–32
Maintenance
Keypad
remove the switch keypad.
on page 6–24.
2. Lift the keypad out of the front case. Refer to Figure 6–23.
CAUTION. Keep the electrical contacts on the back of the keypad clean. Skin oils
and dust will inhibit good electrical contact.
Keypad
Guide
post
Conductive
foam pad
Front case
Figure 6–23: Removing and installing the keypad
6–33
Maintenance
Installation. Use this procedure to install the switch keypad.
1. Align the switch keypad into the front case using the five guide posts as
shown in Figure 6–23.
CAUTION. Keep the electrical contacts on the back of the switch pad clean. Skin
oils and dust will inhibit good electrical contact.
2. Check that all buttons are properly seated into place before proceeding.
3. Place the entire internal assembly into the front case as shown in
Figure 6–16.
instrument.
6–34
Maintenance
Display Shield
STOP. Removing the display shield from the display module will destroy the
display shield gasket. Only remove the shield if you intend to replace it with a
new shield and gasket assembly.
remove the display shield.
on page 6–24.
2. Remove the shield from the display module. Refer to Figure 6–24.
NOTE. You may have to cut through the display shield gasket to remove the
shield. Carefully remove any remaining gasket and adhesive from the LCD
display module.
Display shield
Internal chassis
Front case
Figure 6–24: Removing the display shield
6–35
Maintenance
Installation. Use this procedure to install the display shield.
1. Align the display shield into the front case, using the two guide posts as
shown in Figure 6–25.
2. Remove the paper backing from the adhesive lining of the gasket.
3. Carefully place the entire internal assembly into the front case as shown in
Figure 6–25 making sure that the display module properly contacts the
display shield foam.
instrument.
Internal assembly
Display
shield foam
with
adhesive
Align display shield with
guide
posts on the inside of
the front case.
Displa
y
shield
Guide
post
Front case
Conductive foam pad
(TDS 210 and
TDS 220)
Figure 6–25: Installing the display shield
6–36
Maintenance
Front Case
on page 6–24.
2. Remove the switch keypad using the procedure on page 6–33.
3. Remove (TDS 210 and TDS 220 only) the conductive foam pad from the
front case (see Figure 6–15 on page 6–24).
Installation. Use this procedure to install the front case. Refer to Figure 6–15 on
page 6–24.
1. Place (TDS 210 and TDS 220 only) the conductive foam pad into the front
case.
2. Place the switch keypad into the front case.
3. Place the entire internal assembly into the front case.
instrument.
5. If installing a new front case, place a new front-panel label onto the case as
described on page 6–10.
6–37
Maintenance
Troubleshooting
This section contains information and procedures to help you isolate a defective
module in the oscilloscope.
WARNING. Before doing this or any other procedure in this manual, read the
General Safety Summary and Service Safety Summary found at the beginning of
this manual. Also, to prevent possible injury to service personnel or damage to
electrical components, read Preventing ESD on page 6–1.
Adjustment After Repair
Required Tools and
Equipment
If you replace the main board or power supply, you must adjust the instrument
after repair. Refer to the chapter Adjustment Procedures.
You may need the the following tools and equipment to troubleshoot the
instrument.
Tools and equipment
Oscilloscope with 1X/10X and 100X probes
Tektronix TDS 200- or TDS 300-series with
standard accessory 1X/10X probe and optional
100X probe
DMM
6–38
Example
Tektronix TX-series digital multimeter
Maintenance
Begin Here
Begin with this troubleshooting tree to isolate a problem to a specific troubleshooting procedure.
Plug the instrument into a
proper power source and
press the power button.
Does the
instrument
turn on?
No
Go to
Troubleshooting
Power Supply
(page 6–40).
Yes
Is
the display
functioning
properly?
No
Is there
a display?
No
Go to Checking
Probe Comp
(page 6–40).
Yes
Yes
Check diagnostic screen
displayed at power-on.
Is display
color abnormal
(dark blue, green, or
orange)?
Yes
Go to Troubleshooting
Display (page 6–42).
No
Any
failed
messages?
No
Yes
Go to Checking
Probe Comp
(page 6–40).
Does
backlight
come on?
No
Backlight (page 6–41).
Yes
Instrument should be functioning.
Try the Other Troubleshooting
procedures (page 6–46) and the
Front-Panel Troubleshooting
(page 6–43) if other problems exist.
6–39
Maintenance
PROBE COMP Output
When the instrument is on and operating properly, the PROBE COMP output
should generate a square wave, approximately 5 V in amplitude, at a 1 kHz
frequency. Use the oscilloscope and set the Attenuation switch to 10X on the
P2100 probe to probe this output.
Signal at PROBE COMP
1 kHz signal, no display
Go to Display Troubleshooting (Page 6–42)
No signal
Go to Power Supply Troubleshooting (Page 6–40)
& gt; 1.2 kHz signal
Power Supply
Troubleshooting
Possible problem
Follow these steps to troubleshoot the power supply.
1. Remove the rear case using the procedure Rear Case on page 6–14.
2. Use the test oscilloscope to measure the voltages from the power supply
module at J131 on the main board module. The table below lists the voltages
you should expect to see.
Voltage at J131
Description
Pin 1 (marked with A )
+4.85 VDC to +5.15 VDC
Pin 2
Pin 3
GND
Pin 4
–4.35 VDC to –5.15 VDC (slightly higher if disconnected)
Pin 5
–23 VDC to –25 VDC
Pin 6
GND
Pin 7
1.4 Vpk-pk to 5 Vpk-pk, 47 to 63 Hz square wave
3. If all of the voltages are present, the main board is probably defective.
Replace it.
4. If all or some of the voltages are missing, continue with step 5.
5. Turn off the instrument and disconnect the cable at connector J131 on the
main board module.
6. Turn on the instrument and check the voltages at the loose end of the cable
disconnected from J131, checking for the same voltages as in step 2.
6–40
Maintenance
7. If all of the voltages are missing, check the line fuse on the power supply
module. If the fuse is defective, replace the fuse. If the fuse is ok, continue
with step 8.
8. If all or some of the voltages are missing, the power supply module is
probably defective. Replace it.
Backlight Troubleshooting
Follow these steps to troubleshoot a nonfunctional backlight.
2. Turn the instrument on.
WARNING. To avoid electrical shock, do not touch the two-wire backlight
connector on the power supply module. The output connector is the one closest to
the power line cord.
3. Set the test oscilloscope (with a 100X probe) to measure a 1270 Vpk-pk
(450 VRMS) signal at approximately 60 kHz. Attach the probe ground lead to
the ground lug on the AC line connector and the probe tip to the backlight
connector as shown in Figure 6–26.
AC ground lug
Backlight voltage
test point
Figure 6–26: Measuring the backlight voltage
4. If the 1270 Vpk-pk signal is present, the backlight is probably defective.
Replace the display. If the signal is not present proceed with step 5.
6–41
Maintenance
5. If the 1270 Vpk-pk signal is missing, the power supply module is probably
Follow these steps to troubleshoot a nonfunctional display.
NOTE. Without a functional display module, the simplest way to verify that the
instrument is on and functioning is to check for a signal at the PROBE COMP
output. The PROBE COMP output generates approximately a 5 V, 1 kHz
squarewave.
3. Use the test oscilloscope and set the Attenuation switch to 10X on the P2100
probe to probe the display cable connector at J202 (two channel instruments)
or J102 (four channel instruments) on the main board. The table below
describes the signals you should expect to see.
Signal at J202/J102
Description
–10 VDC to –22 VDC
Pin 2
–23 VDC to –25 VDC
Pin 3
Digital data, high = 5 V, low = 0 V
Pin 4
Pin 5
Pin 6
Pin 7
0 V (ground)
Pin 8
0 V (ground)
Pin 9
+5 VDC
Pin 10
1.89 MHz clock (528 ns period), ≈ 50% duty cycle, appears
in bursts, high = 5 V, low = 0 V
Pin 11
19.3 kHz clock (51.7 ms period), ≈ 16% duty cycle,
Pin 12 (closest to BNC end of
instrument)
75.5 Hz clock (13.2 ms period), ≈ 0.4% duty cycle,
4. If all the signals are present, the display module is probably defective.
Replace it. If all or some of the signals are missing, continue with step 5.
6–42
Maintenance
5. Turn the instrument off and disconnect the display module cable at J202 or
J102 on the main board. Carefully pull the cable straight up from the
connector to avoid damage.
6. Turn the instrument on.
7. Probe the pins of J202 or J102, checking for the same signals as in step 3.
8. If any of the signals are missing, the main board module is probably
Front-Panel
Troubleshooting
Follow these steps to troubleshoot a nonfunctional front panel.
Buttons. Use the manufacturing tests to check that each of the front-panel
controls operate properly.
1. To execute the internal diagnostic tests, remove all input signals from the
instrument and enable the service menus as described on page 5–2.
2. Press the UTILITY button and select Service from the menu selection.
Menu
Settings
Comments
System
Status
Displays the system menus
Do Self Cal
Performs a self calibration
Error Log
Displays a list of any errors logged
This list is useful when contacting a Tektronix Service Center
for help troubleshooting problems
Service
Displays the service menus for adjustment procedures and
diagnostics
3. Select Service Diag. from the menu selection.
Menu
Settings
Comments
FCAL
INIT
Initiates the factory calibration routine to load new calibration
constants
FCAL
STEP
Steps through the calibration routine steps
FCAL
ABORT
Aborts the factory calibration routine without affecting any
Service
Diag.
Displays the service diagnostics menus
6–43
Maintenance
4. Select Mfg. Tests to start the front panel controls test.
Menu
Settings
Comments
Mfg.
Tests
Displays a front panel simulation used to check that all front
panel controls are operational
Loop
Diag.
Displays a menu of choices of how to run the diagnostics
used during manufacturing of the instrument
Repair
Clear
Error
Log
Remove all Error Log entries from memory and set the power
cycle count to zero
5. Select Test FP to start the front panel controls test.
Menu
Test
FP
Settings
Comments
Displays a front panel simulation used to check the operation
of all front panel controls
All buttons must be pressed and all knobs must be turned in
both directions before you can exit this test
Test
Screen
Checks the display screen pixels with different patterns from
each press of the button
Incr.
Contrast
Allows you to check the display contrast operation across
100% of the hardware range
Decr.
Contrast
Stop
Test
Exits the front panel test (Test FP) after all buttons and knobs
have been tested successfully and exits the screen test (Test
Screen)
6. Press each button and rotate each knob to verify the operation of all controls.
If all buttons and knobs have pass the test, pressing the lower menu button
(EXIT) exits the test.
7. If any button fails to operate, inspect and/or clean the switch keypad and the
front panel board contacts. If cleaning does not solve the problem, the switch
keypad, front panel board, or main board may be defective. Use the
procedure below to isolate the problem:
8. Replace the switch keypad and then retest the buttons. If the problem
persists, proceed to step 9.
6–44
Maintenance
9. Use the test oscilloscope and set the Attenuation switch to 10X on the P2100
probe to probe the front panel module cable connector at J603 (two channel
oscilloscopes) or J103 (four channel oscilloscopes) on the main board. The
table below describes the signals you should expect to see.
Signal J603
Description
1
FDCLR
Positive Pulses, 100 ns wide (typical), 280 ms period
2
+5 V
+5 V
3
FPCLR
Negative clock pulses, 100 ns wide (typical), 4.5 ms period
4
GND
0 V (ground)
5
FPDATA
Negative pulse burst, 300 ms period (typical)
Signal J103
Description
1
GND (0 V)
2
Negative clk pulse, 80 ns wide, 4.5 ms period (typical)
3
Positive pulse, 100 ns wide (typical), 280 ms period
4
Positive pulse, 100 ns wide, 280 ms period (typical)
5
Negative pulse, 80 ms wide, 4.5 ms period (typical)
6
GND (0 V)
7
Negative pulse burst, 300 ms (typical)
8
+5 V
9
10
—14 V (typical)
10. If all of the signals are present, the front panel board is probably defective.
Replace it.
11. If some or all of the signals are missing, continue with step 12.
12. Turn the instrument off and disconnect the front panel module cable from the
main board at connector J603 or J103.
13. Turn on the instrument and check connector J603 or J103 for the same
signals as in step 9.
14. If some or all of the signals are missing, the main board is probably
6–45
Maintenance
Diagnostics
The instrument runs an extensive self diagnostics at every power-on. Running the
diagnostics from the service menu will provide no additional information than
the power-up screen and therefore is not needed. The menu selections are only
used during manufacturing of the instrument.
Use these techniques to troubleshoot other parts of the instrument.
Input Connections. Follow these steps to troubleshoot the input connections only
if the following condition is true: the instrument appears to function normally in
every way except you have determined that an input signal is not getting into the
instrument as expected.
1. Remove the rear cover using the procedure Rear Case on page 6–14.
2. Check that the coaxial connections to the back side of the BNC connectors
are intact. Use the DMM to measure continuity from the BNC connector to
the end of the coaxial cable where it attaches to the main board.
3. If no problem is found and the trouble persists, the main board is probably
defective in some way that the internal diagnostics cannot detect. Replace it.
Error Log. Failures that occur during the power-up diagnostics place an entry in
the error log. The error log list can be accessed anytime by pressing the
UTILITY button and selecting Error Log. Table 6–3 lists the possible error codes
and possible causes.
Menu
Settings
Comments
System
Status
Do Self Cal
Error Log
Service
diagnostics
Table 6–3: List of error codes
Error code
Probable cause
0 DI
Diagnostics internal error
Contact Tektronix
4 DI
6–46
Description
CH1 diagnostics failed
Main board or power supply
failure
Maintenance
Table 6–3: List of error codes (cont.)
Error code
Description
Probable cause
5 DI
CH2 diagnostics failed
failure
6 DI
Line trigger diagnostics failed
failure
7 DI
ADG321 diagnostics failed
failure
10 DI
Extension module RAM failed
Extension module failure
11 DI
Extension module checksum
failed
255 Software
Contact Tektronix
Clearing the Error Log. After fixing a problem, clear the error log.
1. To clear the error log, enable the service menus as described on page 5–2.
4. Select Clear Error Log to clear the error log of past failures.
6–47
Maintenance
This section contains the information needed to repackage the oscilloscope for
shipment or storage.
Packaging
Use a corrugated cardboard shipping carton having a test strength of at least
275 pounds (125 kg) and with an inside dimension at least six inches (15.25 cm)
greater than the instrument dimensions.
If the instrument is being shipped to a Tektronix Service Center, enclose the
H
The owner’s address
H
Name and phone number of a contact person
H
Type and serial number of the instrument
H
Reason for returning
H
A complete description of the service required
Seal the shipping carton with an industrial stapler or strapping tape.
Mark the address of the Tektronix Service Center and also your own return
address on the shipping carton in two prominent locations.
Storage
The oscilloscope should be stored in a clean, dry environment. The following
environmental characteristics apply for both shipping and storage (instrument not
operating):
H
Temperature range: –20° C to +60° C
H
Altitude: To 15,000 m
See the chapter Specifications for a complete listing of the environmental
characteristics.
6–48
Options
There are no options available for the TDS 200-Series Digital Real-Time
Oscilloscopes. For a list of available accessories, refer to chapter Mechanical
Parts List.
7–1
Options
7–2
Electrical Parts List
Refer to the Mechanical Parts List chapter for a complete listing and description
of replaceable parts for the TDS 200-Series Digital Real-Time Oscilloscope.
8–1
8–2
Diagrams
See Figure 3–1 on page 3–3 or Figure 3–2 on page 3–4 for a block/interconnect
diagram of your oscilloscope. There are no additional diagrams in this chapter.
9–1
Diagrams
9–2
Replaceable Parts
This chapter contains a list of the replaceable modules for the TDS 200-Series
Digital Real-Time Oscilloscope. Use this list to identify and order replacement
parts.
Parts Ordering Information
Replacement parts are available through your local Tektronix field office or
representative.
Changes to Tektronix products are sometimes made to accommodate improved
components as they become available and to give you the benefit of the latest
improvements. Therefore, when ordering parts, it is important to include the
following information in your order.
H
Part number
H
Instrument type or model number
H
Instrument serial number
H
Instrument modification number, if applicable
If you order a part that has been replaced with a different or improved part, your
local Tektronix field office or representative will contact you concerning any
change in part number.
Change information, if any, is located at the rear of this manual.
Module Servicing
Modules can be serviced by selecting one of the following two options. Contact
your local Tektronix service center or representative for repair assistance.
Module Exchange. In some cases you may exchange your module for a remanufactured module. These modules cost significantly less than new modules and
meet the same factory specifications. For more information about the module
exchange program, call 1-800-833-9200 and select option 2.
New Modules. You may purchase replacement modules in the same way as other
replacement parts.
10–1
Replaceable Parts
Using the Replaceable Parts List
This section contains a list of the mechanical and/or electrical components that
are replaceable for the oscilloscope. Use this list to identify and order replacement parts. The following table describes each column in the parts list.
Table 10–1: Parts list column descriptions
Column
Column name
Description
1
Figure & Index Number
Items in this section are referenced by figure and index numbers to the exploded view illustrations
that follow.
2
Tektronix Part Number
Use this part number when ordering replacement parts from Tektronix.
3 and 4
Serial Number
Column three indicates the serial number at which the part was first effective. Column four
indicates the serial number at which the part was discontinued. No entries indicates the part is
good for all serial numbers.
5
Qty
This indicates the quantity of parts used.
6
Name & Description
An item name is separated from the description by a colon (:). Because of space limitations, an
item name may sometimes appear as incomplete. Use the U.S. Federal Catalog handbook H6-1
for further item name identification.
7
Mfr. Code
This indicates the code of the actual manufacturer of the part.
8
Mfr. Part Number
This indicates the actual manufacturer’s or vendor’s part number.
Abbreviations
10–2
Abbreviations conform to American National Standard ANSI Y1.1–1972.
Replaceable Parts
Mfr. Code to Manufacturer
Cross Index
The table titled Manufacturers cross index shows codes, names, and addresses of
manufacturers or vendors of components listed in the parts list.
Table 10–2: Manufacturers cross index
Mfr.
code
Manufacturer
Address
City, state, zip code
07416
NELSON NAME PLATE COMPANY
3191 CASITAS AVENUE
LOS ANGELES, CA 90039–2410
0KB01
STAUFFER SUPPLY CO
810 SE SHERMAN
PORTLAND, OR 97214–4657
0KB05
NORTH STAR NAMEPLATE INC
5750 NE MOORE COURT
HILLSBORO, OR 97124–6474
0KM03
INSTRUMENT SPECIALTIES CO INC.
505 PORTER WAY
PLACENTIA, CA 92670
14310
AULT INC
7300 BOONE AVE NORTH
BROOKLINE PARK
MINNEAPOLIS, MN 55428
32997
BOURNS INC
TRIMPOT DIVISION
1200 COLUMBIA AVE
RIVERSIDE, CA 92507–2114
77824
SCHLEGEL CORP
1555 JEFFERSON RD
P O BOX 23197
ROCHESTER, NY 14692
80009
TEKTRONIX INC
14200 SW KARL BRAUN DR
PO BOX 500
BEAVERTON, OR 97077–0001
TK0JD
LIGHTWORKS EDITING SYSTEMS LTD
ARIEL HOUSE
74A CHAROLETT ST
LONDON, GB W1P1LR
TK1163
POLYCAST INC
9898 SW TIGARD ST
TIGARD, OR 97223
TK1326
NORTHWEST FOURSLIDE INC
13945 SW GALBREATH DR
SHERWOOD, OR 97140
TK2376
CONDUCTIVE RUBBER TECH
22125 17TH AVE SE, SUITE 117
BOTHELL, WA 98021
TK2569
SHARP CORPORATION
C/O TJBO LIAISON
M/S 78–210
BEAVERTON, OR 97077
TK2647
C/O TEMCO NW
1336 SE 51ST STREET
HILLSBORO, OR 97123
TK6181
IMC PLASTICS INC
19400 SW TETON AVE
TUALATIN, OR 97062
10–3
Replaceable Parts
Table 10–3: Replaceable parts list
Fig. &
index
number
Tektronix part
number
Serial no.
effective
Serial no.
discont’d
Qty
Name & description
Mfr.
code
Mfr. part
number
10-1-1
202–0345–00
B010100
B016657
1
CASE,BACKMACROBLEND DP4–1368,TDS210
TK1163
202034500
202–0345–01
B016658
1
CASE,BACK:BAYBLEND FR110,TDS210
TK1163
202–0345–01
202–0345–00
B010100
1
CASE,BACK:MACROBLEND DP4–1368,TDS220
TK1163
202034500
202–0345–01
B016534
1
CASE,BACK:BAYBLEND FR110,TDS220
TK1163
202–0345–01
B016533
202–0345–01
1
CASE,BACK:BAYBLEND FR110,TDS224
TK1163
202–0345–01
10-1-2
366–0774–00
1
PUSH BUTTON:PC/ABS BAYBLEND,FR110,TEXTURED
TK1163
366–0774–00
10-1-3
213–1099–00
2
SCREW,TPG,PT:K40–1.79 X 31 MM,STEEL,ZINC BLACK
CHROMATE,T–15 TORX– SLOT,FLAT TOP PNH
0KB01
213–1099–00
10-1-4
334–9262–00
1
MARKER,IDENT:LABEL,MARKED WARNING TO AVOID
ELECTRIC,3.765 X 3.990,0.010 POLY
0KB05
334–9262–00
EXTENSION MODULE:OPTIONAL PLUG-IN
ACCESSORY,TDS2CM
80009
TDS2CM
HANDLE,CARRYING:MACROBLEND
TK1163
367–0472–00
10-1-5
10-1-6
367–0472–00
1
10-1-7
348–1484–01
2
FOOT:SANTOPRENE,BLACK,101–64,TEXTURED
TK1163
348–1484–01
10-1-8
159–0399–00
1
FUSE:6.3 AMP,5 X 20 MM,TIME DELAY
TK0JD
159–0399–00
10-1-9
119–5417–00
B010100
C010100
B040351
C010445
1
POWER SUPPLY:11W,5.0V 1.0A,–5.0V 0.3A,–23V
10MA,450VAC 30KHZ 5MA,85–275VAC 45–66HZ
INPUT,TDS210
14310
SW422KA0000F02
119–5417–01
B040352
C010446
B049398
C010878
1
POWER SUPPLY:CUSTOM,AC–DC,11W,85–275VAC 45–66HZ
INPUT,5.0V 1.0A –5.0V 0.3A –23V 10MA 450VAC 30KHZ 5MA
OUTPUT,TDS210
14310
SW422KA0000F02
119–5417–02
B049399
C010879
B089999
C019999
1
OUTPUT,TDS210
14310
SW422KA0000F02
119–5417–00
B010100
C010100
B034184
C010519
1
INPUT,TDS220
14310
SW422KA0000F02
119–5417–01
B034185
C010520
B041057
C011173
1
OUTPUT,TDS220
14310
SW422KA0000F02
119–5417–02
B041058
C011174
B059999
C019999
1
OUTPUT,TDS220
14310
SW422KA0000F02
119–6025–00
B090100
C020100
1
POWER SUPPLY:CUSTOM,12W,5VDC 1.25A,–5VDC 0.7A,
–24VDC 15MA,450 VRMS 6MA 60KHZ,85–275VAC,TDS210
14310
SW422KA0000F03
119–6025–00
B060100
C020100
1
–24VDC 15MA,450 VRMS 6MA 60KHZ,85–275VAC,TDS220
14310
SW422KA0000F03
119–6025–00
1
–24VDC 15MA,450 VRMS 6MA 60KHZ,85–275VAC,TDS224
14310
SW422KA0000F03
441–2082–00
1
CHASSIS ASSY:CKT BD & LCD SUPPORT, MACROBLEND
DP4–1368
TK1163
441–2082–00
10-1-10
10–4
Replaceable Parts
Table 10–3: Replaceable parts list (cont.)
Fig. &
index
number
Tektronix part
number
Serial no.
effective
Serial no.
discont’d
Qty
10-1-11
671–3546–00
B010100
B014281
671–3546–01
B014282
B024456
671–3546–02
B024457
C010100
671–3546–03
Name & description
Mfr.
code
Mfr. part
number
1
CKT BB ASSY:MAIN,TDS210
80009
671–3546–00
1
CIRCUIT BD ASSY:MAIN,TDS210
80009
671–3546–01
B038049
C010495
1
CIRCUIT BD ASSY:MAIN,TESTED,TDS210
80009
671–3546–02
B038050
C010496
B051919
C011104
1
80009
671–3546–03
671–3546–04
B051920
C011105
B089999
C019999
1
80009
671–3546–04
671–4806–00
B090100
C020100
B096042
C020839
1
80009
671–4806–00
671–4806–01
B096043
C020840
B099187
C021678
1
80009
671–4806–01
671–4806–02
B099188
C021679
B103999
C022228
1
80009
671–4806–02
671–4806–03
B104000
C022229
B119999
C029999
1
80009
671–4806–03
671–4806–04
B120000
C030000
1
80009
671–4806–04
671–4024–00
B010100
B014264
1
CIRCUIT BD ASSY:MAIN,TDS220
80009
671–4024–00
671–4024–01
B014265
B022715
1
80009
671–4024–01
671–4024–02
B022716
C010100
B032608
C010567
1
80009
671–4024–02
671–4024–03
B032609
C010568
B042661
C011439
1
CIRCUIT BD ASSY:MAIN,TESTED,TDS220
80009
671–4024–03
671–4024–04
B042662
C011440
B059999
C019999
1
80009
671–4024–04
671–5088–00
B060100
C020100
B063739
C020787
1
80009
671–5088–00
671–5088–01
B063740
C020788
B065809
C021126
1
80009
671–5088–01
671–5088–02
B065810
C021127
B068701
C021265
1
80009
671–5088–02
671–5088–03
B068702
C021266
B079999
C029999
1
80009
671–5088–03
671–5088–04
B080000
C030000
1
80009
671–5088–04
671–4514–01
B010100
B020701
1
CIRCUIT BD ASSY:MAIN BOARD,TESTED,TDS224
80009
671–4514–01
671–4514–02
B020702
B021579
1
80009
671–4514–02
671–4514–03
B021580
1
80009
671–4514–03
10–5
Replaceable Parts
Fig. &
index
number
Tektronix part
number
Serial no.
effective
Serial no.
discont’d
10-1-12
348–1526–00
B010100
C010100
B089999
C019999
348–1526–00
B010100
C010100
B059999
C019999
SHLD,TUBE:COPRESSION,KNITTED WIRE SHIELDING
TUBE,0.312 OD,ACCOM 0.292X0.234 GROOVE, TDS 220
0KM03
8502–0310–40
348–1526–00
B010100
B019999
TUBE,0.312 OD,ACCOM 0.292X0.234 GROOVE, TDS 224
0KM03
8502–0310–40
348–1616–00
B090100
C020100
SHLD GSKT,ELEC FOLD OVER,ADHESIVE,0.25L X
0.51W,BECU BRIGHT NI PLT;TX1/TX3, TDS 210
TK2647 0097–0521–19
348–1616–00
B060100
C020100
0.51W,BECU BRIGHT NI PLT;TX1/TX3, TDS 220
TK2647 0097–0521–19
348–1616–00
B020100
0.51W,BECU BRIGHT NI PLT;TX1/TX3, TDS 224
TK2647 0097–0521–19
10-1-13
Qty
Name & description
1
TUBE,0.312 OD,ACCOM 0.292X0.234 GROOVE, TDS 210
1
Mfr.
code
0KM03
Mfr. part
number
8502–0310–40
CKT BD SUBASSY:OEM FRONT PANEL,3 DETENTED & 4
NON–DETENTED ENCODERS,TDS210/TDS220
32997
3777S–TEK–001
1
CIRCUIT BD ASSY:FRONT PANEL,ENCODER,TDS224
80009
671–4680–00
260–2654–00
1
SWITCH:ELASTOMER KEY PAD,TDS210,TDS220
TK2376 260–2654–00
260–2722–00
10-1-15
1
671–4680–00
10-1-14
679–3743–00
1
SWITCH:PUSHBUTTON SET,ELASTOMERIC,TDS224
TK2376 260–2722–00
B011848
1
LCD SHIELD ASSY:119–4991–00 & 337–4069–00
80009
1
80009
119–4991–11
B012275
1
80009
119–4991–10
119–4991–10
B010100
119–4991–11
B011849
119–4991–10
B010100
119–4991–11
B012276
119–4991–10
1
LCD SHIELD ASSY:119–4991–00,337–4069–00,119–5585–00
80009
119–4991–11
10-1-16
348–1534–00
1
GASKET,EMI:0.375 X 0.50 X 1.0 INCH LONG,OPEN CELL
URETHANE FOAM,W/ADHESIVE STRIP (TDS 210 and
TDS 220 only)
77824
E6819T00100
10-1-17
348–1525–00
2
FOOT:SANTOPRENE,BLACK,64 DEROMETER,TEXTURED
TK1163
348–1525–00
10-1-18
366–0771–00
3
KNOB,CAP:PC/ABS BAYBLEND FR110,SILVER GRAY,0.650
DIA X 0.520 H
TK1163
366–0771–00
10-1-19
366–0772–00
4
KNOB,CAP:PC/ABS BAYBLEND FR110,SIVER GRAY,0.425
DIA X 0.520 H
TK1163
366–0772–00
10-1-20
334–8682–00
1
MARKER,IDENT:FRONT BEZEL LABEL,MARKED
TDS220,5.470 0.345,0.010 POLY FILM, ADEH
0KB05
334–8682–00
334–9261–00
1
TDS210,5.470 X 0.345,0.010 POLY FILM, ADEH
0KB05
334–9261–00
334–9884–00
1
MARKER,IDENT:FRONT BEZEL,MKD TDS224,0.010
POLY,W/ADHESIVE
0KB05
334–9884–00
1
CASE,FRONT:ABS,BAYBLEND,FR110,TDS210
TK1163
202–0346–00
1
TK1163
202–0346–01
1
CASE,FRONT:ABS,BAYBLEND,FR110,TDS220
TK1163
202–0346–00
1
TK1163
202–0346–01
10-1-21
B010100
C010100
202–0346–01
B090100
C020100
202–0346–00
B010100
C010100
202–0346–01
10–6
202–0346–00
B060100
C020100
B089999
C019999
B059999
C019999
Replaceable Parts
Fig. &
index
number
Tektronix part
number
333–4306–00
10-1-22
337–4069–00
10-1-23
367–0476–00
10-1-24
131–6173–00
B010100
131–6173–01
B024107
131–6173–00
B010100
131–6173–01
B022381
Serial no.
effective
Serial no.
discont’d
NOT
SHOWN
119–5585–00
337–4069–00
FLIP STAND:PC/ABS BAYBLEND,FR110,TEXTURED
TK1163
367–0476–00
4
CONTACT,ESD:ESD CLIP,0.005 C17200 BECU,ACCOM 0.062
PCB,0.5 X 0.335 WIDE CONTACT,W/0.230 H X 0.5,3
PRONGED TAB,TDS210
TK1326 131–6173–00
4
PRONGED TAB,TDS210
TK1326 131–6173–00
4
PRONGED TAB,TDS220
TK1326 131–6173–00
4
B012275
07416
PRONGED TAB,TDS220
TK1326 131–6173–00
4
B010100
TK6181 333–4306–00
SHIELD,ELEC:DISPLAY,0.062 GE LEXAN,HARDCOATED
BOTH SURFACES,PKG W/PROTECTIVE SHEETS,
1
131–6173–01
PANEL,FRONT:ABS,TDS224
1
B022380
Mfr.
code
1
B024106
Name & description
PRONGED TAB,TDS224
TK1326 131–6173–00
1
LAMP:CCFT,REPLACEMENT FOR LCD W/LEADS
TK2569 LM0F2964
Qty
Mfr. part
number
10–7
Replaceable Parts
5
4
3
2
1
6
7
8
9
24
23
22
21
10
13
20
11
14
19
15
16
18
12
17
Figure 10–1: Exploded diagram
10–8
Replaceable Parts
Table 10–4 lists the replaceable standard accessories for the oscilloscope and the
instrument options that included them in the original shipment. These accessories
are not shown in Figure 10–1.
Table 10–4: Replaceable standard accessories
Description of accessory
Tektronix part number or nomenclature
Passive 1X/10X probes
P2100
Replacement user manual
English
French
German
Italian
Spanish
Portuguese
Japanese
Korean
Simplified Chinese
Traditional Chinese
Russian
071-0398-XX
071-0400-XX*
071-0402-XX*
071-0401-XX*
071-0399-XX*
071-0403-XX*
071-0405-XX*
071-0408-XX*
071-0406-XX*
071-0407-XX*
071-0404-XX
Replacement power cords
North American
North American
Universal European
United Kingdom
Switzerland
Australian
China
120 V, 60 Hz
240 V, 60 Hz
230 V, 50 Hz
230 V, 50 Hz
230 V, 50 Hz
240 V, 50 Hz
220 V, 50 Hz
161–0230–01
161–0104–08
161–0104–06
161–0104–07
161–0167–00
161–0104–05
161–0304–00
These manuals contain a language overlay for the front-panel controls.
Table 10–5: Replaceable optional accessories
Soft case
AC220
Transit case*
HCTDS32
Measurement Extension Module
TDS2MM
Communications Extension Module
TDS2CM
Rackmount Kit
RM200
Requires the soft case which fits inside.
10–9
Replaceable Parts
Table 10–5: Replaceable optional accessories (cont.)
Replacement extension module instructions
071-0409-XX
071-0483-XX
071-0485-XX
071-0484-XX
071-0482-XX
071-0486-XX
071-0488-XX
071-0491-XX
071-0489-XX
071-0490-XX
071-0487-XX
English
French
German
Italian
Spanish
Portuguese
Japanese
Korean
Simplified Chinese
Traditional Chinese
Russian
Replacement programmer manual
10–10
071-0493-XX